> The Rayleigh criterion is applicable to microscopes and
> Rayleigh did indeed consider this case. Since he considered
> the sample as a collection of self-luminous points (which
> accurately describes a fluorescence image) the condenser does
> not come into it at all. How well the mercury lamp fills the
> pupil of the objective makes no difference to your
> resolution. This is called 'incoherent' imaging because
> light from any one point of the sample cannot interfere with
> light from any other point.
Is this also the reason why there usually is no condenser diaphragm in the
epi-illumination path of the fluorescence microscope? Closing down a
condenser diaphragm would cause parallel and more coherent illumination, but
since fluorescence emission is incoherent it would not change anything to
the resolution or depth-of-field?
Kevin