Hi Jens,
Have you tried redirecting the blue and green channels to the detector that is usually “red”?  
after ~28,000 hrs on our FV-1000, the “green” channel started dropping out, sometimes entirely, other times with dark bands in the scanned frame. The problem persisted when routed to other detectors.  The culprit was the AOTF.

Good luck,
Glen
Glen MacDonald
Digital Microscopy Center
Box 357923
University of Washington
Seattle, WA 98195-7923  USA
(206) 616-4156
[log in to unmask]








> On Jun 1, 2016, at 10:35 AM, jens rietdorf <[log in to unmask]> wrote:
> 
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> Post images on http://www.imgur.com and include the link in your posting.
> *****
> 
> Dear colleagues,
> 
> thank you all for your help on the issue, it turned out a shutter inside
> the scanhead was blocked. This problem was relatively rapidly addressed by
> the service of Olympus Brasil and Olympus Americas. Very helpful input for
> the differential diagnosis came from Farid Jalali of Olympus Canada.
> 
> A new (maybe related) issue occurred now, we experience very low
> sensitivity on detector channels 1&2,
> channel 3 operates normally. the effect appears to be stronger in the
> blue/green part of the spectrum than for the red.
> The FV diagnostic tool does not indicate any abnormalities.
> 
> Again, all ideas and suggestions are very welcome.
> 
> Kind regards, Jens
> 
> Dr. Jens Rietdorf, visiting scientist @ center for technological
> development in health CDTS, Oswaldo Cruz Foundation Fiocruz, Rio de Janeiro
> Brasil.
> 
> On Mon, Apr 4, 2016 at 6:39 PM, jens rietdorf <[log in to unmask]> wrote:
> 
>> Listers,
>> 
>> following the installation of a new air condition, we are not succeeding
>> to get any light after the scanhead of our Olympus FluoView1000 CLSM. End
>> of fiber, there is plenty of light. Behind the scanhead, no light arrives.
>> We work with an inverted IX 81 microscope.
>> 
>> Unfortunately the service situation in Brazil is very poor, so if there
>> were any suggestions how to proceed, I would rather try to check the
>> alignment myself.
>> 
>> Anyone having experience with the FluoView1000 scanhead?
>> 
>> Thank you, Jens
>> 
>> Dr. Jens Rietdorf, visiting scientist @ center for technological
>> development in health CDTS, Oswaldo Cruz Foundation Fiocruz, Rio de Janeiro
>> Brasil.
>> 