Hi Neil,
We have an OMX-V4 including the beam concentration lens for DLM. We do not use DLM very regularly but the few times that we used it it worked ok. Still, the laser power for STORM could be higher…. It really depends on your experiments.
Cheers,
Oliver
Oliver Biehlmaier, PhD | Head of Imaging Core Facility | Biozentrum, University of Basel | Klingelbergstrasse 50/70 | CH-4056 Basel
Phone: +41 61 207 20 73 | Email: [log in to unmask]<mailto:[log in to unmask]> | www.biozentrum.unibas.ch/imcf<http://www.biozentrum.unibas.ch/imcf> | www.microscopynetwork.unibas.ch<http://www.microscopynetwork.unibas.ch>
On 30 Oct 2017, at 15:12, Anthony, Neil <[log in to unmask]<mailto:[log in to unmask]>> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi Ian, thanks for the input.
We're looking at the v4 with the focusing optic as you mentioned.
Can I please ask which version of SoftWoRx you're running, and whether or not you feel that the increased laser density would help the software?
Also, do you know about the z-res? I guess might be very different between v3 and v4 of course, but if it's something you've looked into, I'd appreciate your thoughts.
Thanks
Neil
-----Original Message-----
From: Confocal Microscopy List [mailto:[log in to unmask]] On Behalf Of Ian Dobbie
Sent: Friday, October 27, 2017 2:00 PM
To: [log in to unmask]
Subject: Re: GE DeltaVision OMX DLM
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi Neil
Our experience is slightly unique but might be informative.
We have a v3 OMX with the DLM module. In our experience the power density is too low for good STORM imaging. However the newer OMXs have a lens for focusing the power down on a smaller area. I have no experience if this is adequate or not.
The reconstruction software is however pretty ropey and we generally use something like rainSTORM for localisation imaging post acquisition.
Ian
Sent from my iPhone
On 27 Oct 2017, at 18:12, Anthony, Neil <[log in to unmask]<mailto:[log in to unmask]>> wrote:
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi all,
I was just wondering if anybody has used the single molecule localization add on, DLM, on the GE DeltaVision OMX? If so I'd love to hear your thoughts on resolution, usage, and anything else that struck you.
Thanks in advance for your time.
Neil
________________________________
This e-mail message (including any attachments) is for the sole use of the intended recipient(s) and may contain confidential and privileged information. If the reader of this message is not the intended recipient, you are hereby notified that any dissemination, distribution or copying of this message (including any attachments) is strictly prohibited.
If you have received this message in error, please contact the sender by reply e-mail message and destroy all copies of the original message (including attachments).
|