Hi all,

I think that Vitaly may be telling only part of the story.

1. Using 16 micron pixels is something that you almost have to accept 
if you want back-illumination (Because photoelectrons formed on the 
"back" face of pixels that are much smaller but are 8-9 microns from 
the charge-transfer electrodes and may drift sideways, to be 
collected in neighboring pixels).

The peak QE of such sensors can be >80%, which is good but if you are 
using such a sensor with the electron-multiplier gain turned on so 
that you can count single photons above the background noise level of 
the FET readout amp, then the multiplicative noise of the EM amp 
effectively reduces the actual QE by half (i.e. to 40%. Good, but 
nothing to brag about)

2. The "significant loss of light" is also rather undefined. If you 
use a 2x coupling tube rather than a 1x, on average, each pixel 
SHOULD have only 25% as much signal (assuming the same microscope 
set-up, specimen, focus plane, illumination, and exposure time etc.). 
In fact, it will be a bit less because the extra optics will have 
some reflection losses what may reduce the signal to, say 22%.  If 
you get greater losses, something is wrong. Of course, as S/N depends 
on signal level (in photons, not arbitrary computer units), the 2x 
signal will seem to have more noise because of Poisson Noise, and if 
the signal level is very low (<50 photoelectrons?) then the CCD read 
noise will become more apparent.

However, if you wish to collect high-resolution information, this is 
what you need to do .

3. The comment about confocal  having "0.3 light collection 
efficiency" seems a bit misleading. With a water lens, no more than 
one fluorescent photon in 4 even strikes the glass of the objective. 
Further losses can occur at the dichroic (up to 50%) and the barrier 
filter. So far all of these factors affect widefield CCDs in the same 
way that they affect whatever detector the confocal uses.

It is true that the confocal pinhole CAN be set too small, but if it 
is set to be the same size as the first zero of the back-projected 
Airy pattern, it will pass 80% of the light emitted from the plane of 
focus. This is the signal you want. You may get more photons by 
opening the pinhole, but only as long as there are fluorescent 
substances above or below the plane of focus. Widefield operates as 
though the pinhole is infinitely large. Although, from a 3D object, 
more photons/pixel will strike the detector when using such an 
"Infinite pinhole," the extent to which doing so provides more 
information about the structure of the specimen is not so clear. This 
is another very contentious topic but I think that most will agree 
that photons emerging from objects more than a few microns from the 
plane of focus will contribute more noise than signal.

If leave this topic and limit our consideration to the QE of the 
detectors themselves, at red wavelengths, the effective QE of a 
properly-operated PMT is no more than 10x worse than that of the best 
CCD (i.e., 5x worse than an EM-CCD used in the EM mode). This 
difference is 2x less in the green and less still in the blue. But 
this reduction must be balanced against the fact that out-of-focus 
light, and the Poisson Noise introduced by counting it, is also 
substantially reduced.

With the disk-scanners, one can almost have the best of both worlds: 
Good QE of the EM-CCD as well as the exclusion of far-out-of-focus 
light by a set of pinholes. However, problems still remain in terms 
of the difficulty of changing the effective pinhole size (i.e. 
needing high-NA objectives of different mag), and difficulties in 
spectral imaging.

EM-CCDs with smaller pixels are available now and more are coming. It 
remains to be seen if they their CIC and read-noise performance will 
match those of the present E2V 16 micron chips.

In the mean time, I think that the best plan is to buy a 2x camera 
coupling tube. (and, if you are using this 16 micron sensor in the EM 
mode, take some images with no excitation striking the specimen, just 
to make sure that you aren't picking up a lot of stray light. Keep 
turning off room lights until you only get 2-3 noise pulses per 512 
pixel line.)

Cheers,

Jim P.


>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Dear Michael,
>
>Thanks for the commercial input.
>
>I do not know who is working with the 3.5 micron pixels CCD camera.
>
>However, I have one argument.
>How many labs have an EM-CCD camera with 8 micron pixels? The anwser 
>is rather obvious.
>
>Many labs have Roper's Cascade 512B or Hamamatsu EM-CCD with the 
>"gigantic" pixels.
>
>I tell you it is impossible to avoid undersampling without 
>significant loss of light and working at physiologically relevant 
>concentrations of fluorescently tagged biomolecules.
>
>Anyway, it is known that a (confocal) microscope is a sort of a 
>primitive light-collecting device (ca. 0.3% light collecting 
>efficiency, isn't it?).
>Sensitivity is a serious issue here, but not at the cost of undersampling.
>
>I do hope that Roper's or Hamamatsu sales men would be less bullish 
>and more objective when pushing on 16 um pixel CCDs.
>
>And I do believe that the "gap" in misunderstanding between the 
>Roper's, Hamamatsu, etc. engineers and biologists would narrow down.
>
>Cheers,
>
>Vitaly
>
>NCI-Frederick,
>301-846-6575
>

-- 
               **********************************************
Prof. James B. Pawley,               		            Ph.  608-263-3147 
Room 223, Zoology Research Building,               
FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706  
[log in to unmask]
3D Microscopy of Living Cells Course, June 17-29, 2007, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/	     Applications due by March 15, 2007
	       "If it ain't diffraction, it must be statistics." Anon.