Dear Jennifer,
I don't know any reference but I would do the following quick (and dirty
depending on the applications) approach.
1. threshold the fluoresce and plot it over time or z. You can use on
imagej image->stack->plot z axis profile
2. with that you should get a exponential like decay curve that you can
use as a correction factor for each slice
3. The resulting intensities would be
In=I+I(1-Cn)
where I is corrected intensity for the frame n, I is the real picture
and C the correction factor for the specific frame n.
Please note that you will have worth statistics with higher n since
there is less SNR.
Regards,
NM