Robert,
We have been using DM-NERF for pH measurements (intercellular space in
plants) and it works fine for ex488 & ex514 - dich 527 - em>585. It is
true that this dye doesn't have a true isosbestic point. However, the
emission caused by ex488 is less pH sensitive than that caused by ex514
thus ratio work is possible. We got quite steep and smooth calibartion
curves.
Peter