On 8/2/12 3:57 PM, Pascal Weber wrote:
>
> Leica is working with the company LASOS (Zeiss also) if you type on ther site
> you will not find the reference of the laser. I have a similar problem with the
> same laser and Leica ask me 15,000€ to change it. But the fiber has nothing.
> The laser flash if the power is over 50%. Have you checked if you had good
> power before the fiber at 488nm ?
>
Thanks to many people who replied, both on and off list.
To summarize:
I received a number of replies identifying the QI-Optiq kineFLEX fiber
as the fiber that Leia uses. I also received a suggestion to get teh
fiebr built by Oz Optics.
Concerning the laser, apparently it is possible for the 488 line to lase
into a different mode. I have not yet measured the power in the
individual lines before the AOTF, but I will need to do it. Probably I
will find a prism to disperse the lines so that I can measure them
separately.
--aryeh
--