Hi Jerry
There is no reason why resolution should change except that the rear
aperture is not filled and/or the light does not have the correct degree
of collimation so that sperical aberration may be present. You must
overfill the rear aperture to get good z-resolution -don't focus the
light through the rear aperture. I note that your original figure of z
resultion of 0.2 um for reflected mode sounds very unlikely -that's way
above the diffraction limit.
Cheers Mark
. Jerry Sedgewick wrote:
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi All,
>
> I am working with a graduate student on a home built 2 photon. In the
> past he had routinely used 750nm on a Tsunami (5 watt) and was able to
> see a reflection off the surface of a silicon wafer in order to
> determine the bottom of his sample (the sample itself is polymer beads
> which do not fluoresce at non-polymerizing wavelengths). Now we have
> gone to another laser system (the 10 watt MaiTai)and he is not able to
> see the reflection with the same precision (this probably has nothing
> to do with the new laser, but with spot size in the back aperture or
> something else: it's very well aligned).
>
> Before the new laser (and shorter laser path) he could see the
> reflection with .2 micron accuracy (the reflection fades rapidly with
> .2 micron z-resolution). Now he sees the reflection at about a 2
> micron accuracy.
>
> I have been noticing that the z-resolution in general seems not quite
> up to par with thin focal volumes expected with a multiphoton. This
> new setup seems to have about the same z-resolution as a single
> photon, though I have never measured the actual z-resolution. I don't
> generally use a lens in the light path to converge the beam: when I
> do, it's a 1000mm focal length plano-concave lens which creates a
> narrower diameter spot on the back aperture of the lens. I do use a
> lens to narrow the diameter of the back projected image onto the PMTs.
>
> Any thoughts?
>
> Jerry Sedgewick