Hi all,

We have all seen ray diagrams like.

http://www.antonine-education.co.uk/New_items/MUS/images/ray_diag_6.gif

Light from a point on the axis and in the focal plane diverges until 
some of it hits the (perfect) lens and emerges as a parallel ray 
bundle on the other side. However, things become a little more 
complex when the focal-plane emitter is moved off the axis, as in

http://www.phg.ulg.ac.be/Hololab/fichiers/Graphic1.jpg

Now the emergent parallel ray bundle is at an angle to the axis. So 
you would think that, as long at the next element in the optical 
system (the tube lens) was sufficiently large enough (or close 
enough) to collect the angled ray bundle created by light emerging 
from the focus plane at the edge of the field of view, then all the 
light from the field of view would be focused into the Intermediate 
Image. Then it would pass through the eyepiece/camera-coupler and end 
up in the recorded image.

But what about light emerging from above or below the focal plane? 
Although I couldn't find an on-line sketch, it is clear that instead 
of producing a parallel emerging ray bundle, it will produce one that 
either diverges or converges (resp). So some of these rays, 
particularly those from points both off axis and out of focus, will 
leave the objective at angles even larger than those from the edge of 
the field of view at the plane of focus. How much more will depend on 
how far off axis and away from the focal plane they are, but when one 
focuses tens of micrometers into a thick, fluorescent specimen, it 
can be quite a large angle.

As we want to place reasonable limits on the diameter of the black 
metal tube between the objective and the tube lens, (and on the 
diameter of the tube lens itself), then compromises must be made and 
some of the light from out-of-focus, off-axis sources will strike the 
wall and be lost. The general term for this loss is "Vignetting".

I know nothing about the actual changes that Olympus has apparently 
incorporated in their LV200 except what I found at

http://www.olympus-europa.com/corporate/1696_1948.htm

However, as I remember it, bioluminescence has a very low signal and 
the specimens are often quite 3-dimensional. To the extent that this 
is true and that a high NA objective is used to collect the signal, 
much of this signal may be out of focus. So I guess that having a 
tube lens with a large diameter will collect more of this light . Of 
course, it won't focus this extra light very well (the out of focus 
sources will still look like pale, diffuse blobs. You can't get 
around this.) but the total number of photons in the image will be 
larger with this large tube lens than with a "normal" one.

Two last notes:

1) If the tube lens has a larger diameter, the cone of in-focus-light 
that converges from it to focus at the image plane will have a larger 
half angle, and hence a higher NA. However, unless the magnification 
of the entire system is very low indeed (i.e. using a 1-3x infinity 
objective), then this tube NA has no affect on the sharpness of the 
final image (unless the large NA tube lens is not properly corrected 
to produce near-diffraction-limited performance at this NA: in which 
case the final image may be less sharp).

2) To some extent, all microscopes have this limitation (off-axis, 
out-of-focus light is preferentially lost) but it will be 
particularly severe when using high-NA objectives of lower 
magnification because these have larger fields of view. Indeed, 
vignetting and other limitations on the performance of objectives are 
discussed (with better figures!) in Chapters 7 and 11 of the Handbook 
of Biological Confocal Microscopy.

Cheers,

Jim Pawley
-- 
               **********************************************
Prof. James B. Pawley,               		            Ph.  608-263-3147 
Room 223, Zoology Research Building,               
FAX  608-265-5315
1117 Johnson Ave., Madison, WI, 53706  
[log in to unmask]
3D Microscopy of Living Cells Course, June 13-25, 2009, UBC, Vancouver Canada
Info: http://www.3dcourse.ubc.ca/	     Applications due by March 15, 2009
	       "If it ain't diffraction, it must be statistics." Anon.