On 8/20/12 4:45 AM, George McNamara wrote:
>
> If this is correct, then you are using the scanning mirrors to create
> the image. Since you are not using a point illumination source (the
> focused lasers), you will probably be better off opening the pinhole to
> collect more light. However, on the spectral confocal systems I am
> familiar with (Zeiss LSM710 and Leica SP5 and SP1) opening the pinhole
> compromises the spectral resolution.
>
In the Nikon C1 and A1 spectral, the spectral unit is on the other side
of a fiber, while the pinhole is in the scanhead. Therefore, spectral
resolution should be independent of pinhole size.
--aryeh
--
Aryeh Weiss
Faculty of Engineering
Bar Ilan University
Ramat Gan 52900 Israel
Ph: 972-3-5317638
FAX: 972-3-7384051
|
|
|