Kevin,
	I'll admit, I have no idea what a notch filter is. Would you 
have a moment to explain the basis of its function? I replied to you 
on the list in case there are few others also curious.

	Thanks,
		Tobias

>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Okay, when hitting the send button I just realized my mistake. The notch
>filter in the epifluo cube will not work on the confocal since the
>excitation path coincides with the emission path at that point. Sorry for
>that. It is different on our widefield setup.
>
>On the confocal you would have to put the notch filter on the emission side
>of the first dichroic. Generally you won't have access to that very easily.
>I guess you will have to wait for the manufacturers after all ...
>
>Best regards,
>
>Kevin
>
>
>Kevin Braeckmans, Ph.D.
>Lab. General Biochemistry & Physical Pharmacy
>Ghent University
>Harelbekestraat 72
>9000 Ghent
>Belgium
>Tel: +32 (0)9 264.80.78
>Fax: +32 (0)9 264.81.89
>E-mail: [log in to unmask]
>
>
>>  -----Oorspronkelijk bericht-----
>>  Van: Confocal Microscopy List
>>  [mailto:[log in to unmask]] Namens Reece, Jeff
>>  (NIH/NIEHS) [E]
>>  Verzonden: dinsdag 5 september 2006 22:38
>>  Aan: [log in to unmask]
>>  Onderwerp: Re: High NA objectives for confoal microscopy
>>
>>  Search the CONFOCAL archive at
>>  http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>>  I'll start with a minor point.  The RI of cytoplasm has been
>>  reported in the range 1.35-1.37.  So if you use RI=1.36
>>  instead of 1.33, the critical angle is now 63 degrees. On a
>>  confocal microscope, the intensity of excitation light within
>>  the solid angle between 63 and 67 degrees is much less than
>>  that within 0-63 degrees.  I would guess the proportion
>>  varies wildly from manufacturer to manufacturer, microscope
>>  to microscope within a manufacturer, and possibly even laser
>>  to laser on a single microscope.  I wouldn't trust any
>>  theoretical derivations of this.
>>
>>  On the other hand, when you are talking about excitation at
>>  z=0, the coverglass surface, from angles close to the
>>  critical angle, the TIRF excitation is about 4-5x more
>>  efficient than the non-TIRF excitation (according to
>>  Axelrod's 1984 paper; TIRF users might have noticed this
>>  non-intuitive effect).  So, it is a good question, especially
>>  if you are trying to justify the purchase of a $8-10k water
>>  immersion lens: how much of the confocal signal at z=0 is
>>  actually due to TIRF?  You should be able to get a decent
>>  measure of this if you have NA=1.4 and NA=1.3 lenses, looking
>>  at two focal planes: z=0 and a couple of microns further into
>>  the sample.  The sample ideally would have sparse, punctate
>>  fluorescence in both focal planes.
>>
>>  A related question: for most microscope users with fixed
>>  samples, does the RI mismatch problem go away?  I know that
>>  the subject of mounting media has been discussed before on
>>  this listserv, but I haven't found from that discussion or
>>  anywhere else a mountant that prevents photobleaching, has
>>  RI=1.52, solidifies, and does not require dehydration of the
>>  sample.  Please everyone, let me know if such a mountant exists. 
>>
>>  Now to one of my current pet peeves, but I think it's an
>>  important tangent.  I think the biggest problem for most
>>  confocal users using
>>  NA=1.4 and a sample of unmatched RI is not spherical
>>  aberration or TIRF signal. It is the sometimes significant
>>  reflection that gets back to the detector when near the cover
>>  glass, especially with low fluorescence signal.  This
>>  situation does not go away if everyone uses the perfect
>>  mounting medium on fixed cells.  Many confocal users with
>>  live cells are interested only in the region just above the
>>  coverglass, and of course would rather use the NA=1.4 lens
>>  because of its better resolution and more efficient light
>>  collection.  Having TIRF going on at the same time could
>  > actually be advantageous. In such cases, the filters used by
>>  the manufacturers do not provide enough blocking of the
>>  excitation light.
>>  My hope is, for the practical benefit of this significant
>>  group of users, that the microscope designers have already
>>  started addressing this issue, taking advantage of recent
>>  advances in filter technology.  I propose that, with NA=1.4
>>  and detector gain at its maximum, one should not be able to
>>  detect the glass-water interface with an XZ scan.
>>  Accomplishing this should require a relatively inexpensive
>>  upgrade to most existing CLSM systems, as well as adding
>>  little to the production cost of new systems.  If I'm wrong
>>  about the cost, let it be an option.
>>
>>  Cheers,
>>  Jeff
>>
>>  Jeff M. Reece
>>  Confocal Microscopy Center
>>  National Institute of Environmental Health Sciences (NIEHS)
>>  National Institutes of Health (NIH) P.O. Box 12233, MD F2-02
>>  Research Triangle Park, NC  27709
>>  (919) 541-0311
>>  [log in to unmask]
>>
>>
>>
>>  > -----Original Message-----
>>  > From: Haridas Pudavar [mailto:[log in to unmask]]
>>  > Sent: Sunday, September 03, 2006 5:24 PM
>>  > To: [log in to unmask]
>>  > Subject: High NA objectives for confoal microscopy
>>  >
>>  > ---------------------- Information from the mail header
>>  > -----------------------
>>  > Sender:       Confocal Microscopy List
>>  <[log in to unmask]>
>>  > Poster:       Haridas Pudavar <[log in to unmask]>
>>  > Subject:      High NA objectives for confoal microscopy
>>  > --------------------------------------------------------------
>>  > -----------------
>>  >
>>  > This is a multi-part message in MIME format.
>>  > --------------020807090008000002070803
>>  > Content-Type: text/plain; charset=ISO-8859-1; format=flowed
>>  > Content-Transfer-Encoding: 7bit
>>  >
>>  > Search the CONFOCAL archive at
>>  > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>  >
>>  > Hi,
>>  > I was wondering what is the highest NA objective one can use for
>>  > confocal without getting into total internal reflection at
>>  > coverslip/media interface. It seems to me that with NA 1.4,
>>  the angle
>>  > of incidence (~67 degree) is already higher than critical
>>  angle. (~ 61
>>  > degree assuming a RI of 1.52 for oil and 1.33 for cell/media.)
>>  > Therefore, NA above 1.3  (~ 57 degree  for 1.3, which is
>>  smaller than
>>  > critical angle)will give huge TIR from coverslip/media interface.
>>  >  But all the microscopy manufacturers are selling 1.4 NA objectives
>>  > for high resolution confocal microscopy.
>>  > Am I making any fundamental mistake in the calculation ?
>>  > Can somebody help me?
>>  >
>>  >  Haridas
>>  >
>>  >
>>  >
>>  > --
>>  > Haridas Pudavar, Ph.D.
>>  > Research Asst. Professor,
>>  > Inst. for Lasers, Photonics & BioPhotonics,
>>  > 458, NSC, Dept. of Chemistry,
>>  > SUNY at Buffalo, Buffalo, NY-14260
>>  > Ph: (716) 645 6800 x  2220
>>  > Fax: (716) 645 6945
>>  >
>>  >
>>  > --------------020807090008000002070803
>>  > Content-Type: text/html; charset=ISO-8859-1
>>  > Content-Transfer-Encoding: 7bit
>>  >
>>  > Search the CONFOCAL archive at
>>  > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>  > <!DOCTYPE html PUBLIC "-//W3C//DTD HTML 4.01 Transitional//EN">
>>  > <html>
>>  > <head>
>>  >   <meta content="text/html;charset=ISO-8859-1"
>>  > http-equiv="Content-Type">
>>  > </head>
>>  > <body bgcolor="#ffffff" text="#000000">
>>  > <p class="MsoNormal">Hi,<br>
>>  > I was wondering what is the highest NA objective one can use
>>  > for confocal without getting into total internal reflection at
>>  > coverslip/media
>>  > interface. It seems to me that with NA 1.4,<span style="">&nbsp;
>>  > </span>the angle of incidence (~67 degree) is already higher than
>>  > critical
>>  > angle. (~ 61 degree assuming a RI of 1.52 for oil and 1.33 for
>>  > cell/media.)<span style="">&nbsp; </span>Therefore, NA
>>  above 1.3&nbsp;
>>  > (~ 57 degree&nbsp; for 1.3, which is smaller than critical
>>  > angle)will give
>>  > huge TIR from coverslip/media interface. <span style=""><br>
>>  > &nbsp;</span>But all the microscopy manufacturers are
>>  > selling 1.4 NA objectives for high resolution confocal
>>  microscopy.<br>
>  > > <span style=""></span>Am I making any fundamental mistake in the
>>  > calculation
>>  > ?<span style="">&nbsp; </span><br>
>>  > Can somebody help me?</p>
>>  > <p class="MsoNormal"><o:p>&nbsp;Haridas<br>
>>  > <br>
>>  > </o:p></p>
>>  > <br>
>>  > <br>
>>  > <pre class="moz-signature" cols="72">--
>>  > Haridas Pudavar, Ph.D.
>>  > Research Asst. Professor,
>>  > Inst. for Lasers, Photonics &amp; BioPhotonics,
>>  > 458, NSC, Dept. of Chemistry,
>>  > SUNY at Buffalo, Buffalo, NY-14260
>>  > Ph: (716) 645 6800 x  2220
>>  > Fax: (716) 645 6945</pre>
>>  > </body>
>>  > </html>
>>  >
>>  > --------------020807090008000002070803--
>>  >
>>


-- 
       _      ____          __   ____   
      /  \   /          / \    /   \ \        Tobias I. Baskin
     /   /  /          /   \   \      \         Biology Department
    /_ /   __      /__ \   \       \__    611 N. Pleasant St.
   /      /          /       \   \       \        University of Massachusetts
  /      /          /         \   \       \	    Amherst, MA, 01003
/      / ___   /           \   \__/  \ ____
http://www.bio.umass.edu/biology/baskin/
Voice: 413 - 545 - 1533 Fax: 413 - 545 - 3243

ATOM RSS1 RSS2