You make some interesting points!  It does seem that the low threshold
capability of PMT's make them more useful for our photon counting type
experiments.  Still, in the case of a brighter sample the speed of an
EMCCD would be useful.

Thanks,

Craig

On 10/6/07, George McNamara <[log in to unmask]> wrote:
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>  Hi Craig,
>
>  The Lightform PARISS has a detection slit in front of the spectral
> dispersion element and the CCD. This makes it half-confocal. If a matching
> slit is placed in the epi-illumination field aperture plane (or a conjugate
> plane) it is a slit confocal (aka line scan confocal). The result is an
> X*Spectral image. If the microscope has a motorized stage, the Y-axis can be
> added = push broom confocal imaging spectrometer. The performance is such
> that the emission filter can be removed from the epi-illumination filter
> cube when used with an Hg lamp (the bright Hg line or lines are one
> nanometer or so side = 1 or a few pixels, so do not bleed into the emission
> areas and can be used to quantify the excitation power). The PARISS can also
> be used for brightfield images.
>
>  You can contact Jeremy Lerner at http://www.lightforminc.com/ for
> additional details about the PARISS.
>
>
>  With respect to PMT vs CCD or EMCCD, Pawley's rule is that a typical point
> scanning confocal microscope detects a maximum of about 8 photons per pixel
> per scan. This is a nice match to a PMT (or APD for far red and infrared
> photons), not so good for a CCD (see Jim's post about noise). On the other
> hand, a CCD is an area detector (or line * spectrum detector for the
> PARISS), so the CCD is a great way to acquire 2D images fast. A CCD or EMCCD
> system can be used to image a thin optical section, quickly, with judicious
> use of excitation light, as in TIRF, fast multipoint multiphoton scanners
> such as the TriMScope (if it can be made to work reliably) or Scherer's
> stochastic scanner (
> http://www.opticsexpress.org/abstract.cfm?id=89328), or
> hopefully by widefield multiphoton excitation (Brooker 2007 US patent
> 7,170,675 ), or by multiplane acquisition followed by digital deconvolution.
> If you can buy or borrow 3 EMCCD's like Ian Parker, you can acquire multiple
> focal planes simultaneously, see PubMed 17716727, also similar rig by
> Prabhat, PubMed 17384151.
>
>  Back to 32-channel PMT confocal microscopes: the Zeiss META detector
> includes blockers for the major laser lines. As far as I am aware, they are
> in the light path all the time (the 488 nm blocker may be responsible for
> zigzag artifacts in CFP emission spectra, though bad PMT channels could also
> be responsible). Nikon may include similar blockers in the si. The Zeiss
> LSM510 light path can be configured to use a microscope filter cube. With
> the right cube and a mirror slide, you can acquire the spectrum of the Hg
> lamp using the META, or you could buy a Lightform MIDL like Bob Zucker's and
> just lay it on the microscope. The META spectrum of the MIDL lamp is not
> even close to the calibration curve that comes with the MIDL. Bob and Jeremy
> have published the close match of the PARISS data for the MIDL.
>
>
>  Enjoy,
>
>
>  George
>
>
>
>  At 07:26 PM 10/5/2007, you wrote:
>
> Search the CONFOCAL archive at
>  http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>  Thanks!  We are interested in co-localization, for which we sometimes
>  need the third dimension.  Mainly we want to do spectral unmixing for
>  multiple dyes and colocalization, so we really do need optical
>  slicing!
>
>  Thanks,
>
>  Craig
>
>
>  On 10/5/07, Robert Zucker <[log in to unmask]> wrote:
>  > Search the CONFOCAL archive at
>  > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>  >
>  > Craig
>  > you are correct.
>  > Currently the PARISS  only detects X-Y--not depth.
>  > however, most data that I have seen from  confocal spectral systems are
>  > only a single XY slice.  they usually are not used to do a 3D
>  > reconstruction in spectral mode.
>  > bob
>  > .
>  >
>  >
>  > Robert M. Zucker, PhD
>  > U.S. Environmental Protection Agency
>  > Office of Research and Development
>  > National Health and Environmental Effects Research Laboratory.
>  > Telephone: 919-541-1585   Fax: 919-541-4017
>  > e-mail: [log in to unmask]
>  >
>  > Mail address: Reproductive Toxicology Division, MD 67
>  > 2525 E.NC Highway 54
>  > Research Triangle Park, North Carolina, 27711
>  >
>  > Shipping address: 2525 E.NC Highway 54
>  > Durham, NC, 27713
>  >
>  >
>  >
>  >
>  >              Craig Brideau
>  >              <craig.brideau@G
>  >              MAIL.COM>
>             To
>  >              Sent by:                 [log in to unmask]
>  >              Confocal
>             cc
>  >              Microscopy List
>  >              <CONFOCAL@LISTSE
>      Subject
>  >              RV.BUFFALO.EDU>          Re: Nikon C1si?
>  >
>  >
>  >              10/05/2007 06:44
>  >              PM
>  >
>  >
>  >               Please respond
>  >                     to
>  >                  Confocal
>  >              Microscopy List
>  >              <CONFOCAL@LISTSE
>  >              RV.BUFFALO.EDU>
>  >
>  >
>  >
>  >
>  >
>  >
>  > Search the CONFOCAL archive at
>  > http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>  >
>  > It doesn't seem to deal with depth though?  It only detects in X-Y, so
>  > can it generate optical slices?
>  >
>  > Craig
>  >
>  > On 10/5/07, Robert Zucker <[log in to unmask]> wrote:
>  > > Search the CONFOCAL archive at
>  > >
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>  > >
>  > > The PARISS system uses a Retiga 2000R cooled CCD camera. It is far
>  > more
>  > > sensitive and accurate than any confocal microscope.
>  > >  You can detect things in that system that can not be observed with a
>  > > confocal spectral system.  It is a great system in my opinion.
>  > > Bob
>  > >
>  > > Robert M. Zucker, PhD
>  > > U.S. Environmental Protection Agency
>  > > Office of Research and Development
>  > > National Health and Environmental Effects Research Laboratory.
>  > > Telephone: 919-541-1585   Fax: 919-541-4017
>  > > e-mail: [log in to unmask]
>  > >
>  > > Mail address: Reproductive Toxicology Division, MD 67
>  > > 2525 E.NC Highway 54
>  > > Research Triangle Park, North Carolina, 27711
>  > >
>  > > Shipping address: 2525 E.NC Highway 54
>  > > Durham, NC, 27713
>  > >
>  > >
>  > >
>  > >
>  > >              Craig Brideau
>  > >              <craig.brideau@G
>  > >              MAIL.COM>
>  > To
>  > >              Sent by:                 [log in to unmask]
>  > >              Confocal
>  > cc
>  > >              Microscopy List
>  > >              <CONFOCAL@LISTSE
>  > Subject
>  > >              RV.BUFFALO.EDU>          Re: Nikon C1si?
>  > >
>  > >
>  > >              10/05/2007 06:19
>  > >              PM
>  > >
>  > >
>  > >               Please respond
>  > >                     to
>  > >                  Confocal
>  > >              Microscopy List
>  > >              <CONFOCAL@LISTSE
>  > >              RV.BUFFALO.EDU>
>  > >
>  > >
>  > >
>  > >
>  > >
>  > >
>  > > Search the CONFOCAL archive at
>  > >
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>  > >
>  > > Will this system have the same sort of sensitivity as a multi-anode
>  > > PMT system like the C1si?  This PARISS system seems to use a CCD
>  > > camera as its detector, which will not be as sensitive, if I
>  > > understand correctly?
>  > >
>  > > Craig
>  > >
>  > > On 10/5/07, Robert Zucker <[log in to unmask]> wrote:
>  > > > Search the CONFOCAL archive at
>  > > >
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>  > > >
>  > > > Criag
>  > > > If spectral analysis is the endpoint you may want to consider the
>  > > PARISS
>  > > > system from Lightform. ( http://www.lightforminc.com/) that can be
>  > > > placed on a widefield microscope.
>  > > > We are using the PARISS system to obtain spectral data --it has 1nm
>  > > > resolution and has a spectrum from 400-900nm. I find it to be a
>  > great
>  > > > asset in the laboratory,
>  > > > It is far more sensitive and accurate that any confocal system that
>  > I
>  > > > have seen.
>  > > >
>  > > > Contact Jeremy Lerner for details on his system.
>  > > > best wishes
>  > > > Bob
>  > > >
>  > > >
>  > > >
>  > > >
>  > > >
>  > > >
>  > > >
>  > > >
>  > > >                                 LightForm, Inc.,
>  > > >                           601 Route 206, Suite 26-479
>  > > >                              Hillsborough NJ 08844
>  > > >                                Tel: (908)281 9098
>  > > >                         Email: [log in to unmask]
>  > > >
>  > > >
>  > > >
>  > > >
>  > > >
>  > > >
>  > > > .
>  > > >
>  > > > Robert M. Zucker, PhD
>  > > > U.S. Environmental Protection Agency
>  > > > Office of Research and Development
>  > > > National Health and Environmental Effects Research Laboratory.
>  > > > Telephone: 919-541-1585   Fax: 919-541-4017
>  > > > e-mail: [log in to unmask]
>  > > >
>  > > > Mail address: Reproductive Toxicology Division, MD 67
>  > > > 2525 E.NC Highway 54
>  > > > Research Triangle Park, North Carolina, 27711
>  > > >
>  > > > Shipping address: 2525 E.NC Highway 54
>  > > > Durham, NC, 27713
>  > > >
>  > > >
>  > > >
>  > > >
>  > > >              Craig Brideau
>  > > >              <craig.brideau@G
>  > > >              MAIL.COM>
>  > > To
>  > > >              Sent by:                 [log in to unmask]
>  > > >              Confocal
>  > > cc
>  > > >              Microscopy List
>  > > >              <CONFOCAL@LISTSE
>  > > Subject
>  > > >              RV.BUFFALO.EDU>          Re: Nikon C1si?
>  > > >
>  > > >
>  > > >              10/05/2007 05:48
>  > > >              PM
>  > > >
>  > > >
>  > > >               Please respond
>  > > >                     to
>  > > >                  Confocal
>  > > >              Microscopy List
>  > > >              <CONFOCAL@LISTSE
>  > > >              RV.BUFFALO.EDU>
>  > > >
>  > > >
>  > > >
>  > > >
>  > > >
>  > > >
>  > > > Search the CONFOCAL archive at
>  > > >
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>  > > >
>  > > > Thanks for your input!  As in everything, there's a trade-off.
>  > Still,
>  > > > for us the spectral data will be worth it.
>  > > >
>  > > > Craig
>  > > >
>  > > > On 10/5/07, Robert Zucker <[log in to unmask]> wrote:
>  > > > > Search the CONFOCAL archive at
>  > > > >
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>  > > > >
>  > > > > Craig
>  > > > > The efficiency of detection is much less in the spectral mode due
>  > to
>  > > > the
>  > > > > design of the multianode detector. In addition as Kurt Thorn said
>  > > the
>  > > > > light is split into a number of different bandpass channels
>  > instead
>  > > of
>  > > > > being detected by only one channel which also limits the detection
>  > > in
>  > > > > each channel.
>  > > > > Nikon compensates for this decreased light by increasing the pixel
>  > > > dwell
>  > > > > time. However photons are photons. With less photons you will get
>  > a
>  > > > > noisier image.
>  > > > > The spectral detection of any confocal system using a multianode
>  > > > > detector will not be as good as a PMT designed for good
>  > sensitivity
>  > > > and
>  > > > > low light detection. It will produce nosier images but if there is
>  > > > > enough light you will be able to determine a valuable spectrum
>  > which
>  > > > can
>  > > > > be used for scientific experiments and to manipulate images. Like
>  > > > > everything with confocal microscopy there is trade-offs and no
>  > > perfect
>  > > > > system.
>  > > > > Best wishes.
>  > > > > Bob
>  > > > >
>  > > > > Robert M. Zucker, PhD
>  > > > > U.S. Environmental Protection Agency
>  > > > > Office of Research and Development
>  > > > > National Health and Environmental Effects Research Laboratory.
>  > > > > Telephone: 919-541-1585   Fax: 919-541-4017
>  > > > > e-mail: [log in to unmask]
>  > > > >
>  > > > > Mail address: Reproductive Toxicology Division, MD 67
>  > > > > 2525 E.NC Highway 54
>  > > > > Research Triangle Park, North Carolina, 27711
>  > > > >
>  > > > > Shipping address: 2525 E.NC Highway 54
>  > > > > Durham, NC, 27713
>  > > > >
>  > > > >
>  > > > >
>  > > > >
>  > > > >              Craig Brideau
>  > > > >              <craig.brideau@G
>  > > > >              MAIL.COM>
>  > > > To
>  > > > >              Sent by:
>  > [log in to unmask]
>  > > > >              Confocal
>  > > > cc
>  > > > >              Microscopy List
>  > > > >              <CONFOCAL@LISTSE
>  > > > Subject
>  > > > >              RV.BUFFALO.EDU>          Re: Nikon C1si?
>  > > > >
>  > > > >
>  > > > >              10/04/2007 02:49
>  > > > >              PM
>  > > > >
>  > > > >
>  > > > >               Please respond
>  > > > >                     to
>  > > > >                  Confocal
>  > > > >              Microscopy List
>  > > > >              <CONFOCAL@LISTSE
>  > > > >              RV.BUFFALO.EDU>
>  > > > >
>  > > > >
>  > > > >
>  > > > >
>  > > > >
>  > > > >
>  > > > > Search the CONFOCAL archive at
>  > > > >
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>  > > > >
>  > > > > I was wondering if you saw this decreased throughput in spectral
>  > > mode
>  > > > > only, or also in conventional detection mode?  The spectral mode
>  > > works
>  > > > > with a multi-anode PMT, which has a lower quantum efficiency than
>  > a
>  > > > > conventional PMT.  From what I have seen of the design, it also
>  > has
>  > > > > the option to use a conventional PMT system with the device, and I
>  > > am
>  > > > > wondering if you have also tried that mode of operation?
>  > > > >
>  > > > > Thanks,
>  > > > >
>  > > > > Craig
>  > > > >
>  > > > > On 10/4/07, Robert Zucker <[log in to unmask]> wrote:
>  > > > > > One of the major limitations that we have seen with this unit
>  > and
>  > > > also
>  > > > > > with  the Zeiss meta 510 is the decreased light throughput .
>  > This
>  > > > > > creates images that are noisier than conventional confocal
>  > > > > microscopes.
>  > > > > > In our hands it appears you will need to have a bright sample to
>  > > > make
>  > > > > > the spectral system work properly. ,
>  > > > >
>  > > >
>  > >
>  >
>
>
>
>
>
>
>  George McNamara, Ph.D.
>  University of Miami, Miller School of Medicine
>  Image Core
>  Miami, FL 33010
>  [log in to unmask]
>  [log in to unmask]
>  305-243-8436 office
>  http://home.earthlink.net/~pubspectra/
>  http://home.earthlink.net/~geomcnamara/
> http://www.sylvester.org/health_pro/shared_resources/index.asp
> (see Analytical Imaging Core Facility)
>
>
>