Hi Olaf: Many materials have coefficient of thermal expansion on the
order of 10^(-5) 1/K. So if your microscope was a solid block and the
distance between the objective and the stage was 1 cm, then you could
expect variation on the order of 0.1 micron per degree. I imagine that
variations in the refractive index of immersion oil with temperature
may have a greater effect on focusing.
Michael Model
Confocal Imaging Core,
Department of Biological Sciences,
Kent State University,
Kent, OH 44242
tel. 330-672-2874
----- Original Message -----
From: "Dr. Olaf Selchow" <[log in to unmask]>
Date: Tuesday, November 14, 2006 3:29 pm
Subject: Thermal Focus Drift - Olympus BX 51 WI -
To: [log in to unmask]
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hello List,
>
> anybody out there who has an idea about how much focus shift
> (approximately) I can expect from a 5 (10) deg celsius temperature
> change in the lab on an Olympus BX 51 WI?
> With aproximately I mean: is it 1, 5 or 10 µm?? Or even more?
>
> I know I should keep the temperature constant (the goal in our
> case is
> plus minus 1 deg C) but we need to know how much the focus shifts
> to
> have good arguments to ask for better air conditioning the lab...
> A
> measurement would take time - therefore it would help a lot if
> someone
> has done the test already and can help out with his experience.
>
> Thanks very much!
>
> Olaf
>
|
|
|