Michael,
>We are planning the configuration for a multi-photon microscope. Does
>anybody have knowledge/opinions about
>1. IR coating/correction in Nikon vs. Olympus optics; and
Can't answer; don't know
>2. PSF (obviously only important in this application for delivery of
>excitation) in Nikon vs. Olympus optics.
>Thanks-
MP or not, get vendor representatives to send you as many of the same
objectives as possible and using a reference sample (e.g., 100 nm
fluorescent beads), determine the PSF and brightness (coating
effects). No two objectives from the same manufacturer, let alone
different ones, will be identical. In fact, if budget is a
significant consideration, ask for a better discount on the lens of
lesser quality as long as they meet your needs.
Good luck
>
>____________________________________________________________________________
>Michael Cammer Analytical Imaging Facility Albert Einstein Coll. of Med.
>1300 Morris Park Ave. Bronx, NY 10461 (718) 430-2890 Fax: 430-8996
>URL: http://www.aecom.yu.edu/aif/
>personal: http://cammer.home.mindspring.com/
--
Mario M. Moronne, Ph.D.
NanoMed Technologies
ph (510) 528-2400
Berkeley, CA
94706
|
|
|