Shading can arise from excitation or emission detection problems. As I understand it, if the former it is due to the rear aperture of the objective not being filled with a uniform field during scans. This could arise from inadequate laser beam expansion and/or mislaligment induced by the scan system. Poorly designed scan systems could also induce shading on the emission side.
Perhaps you might look at the rear aperture illumination alignment during slow scans to decide where the problem lies? It should be quite uniform and not chnage during scanning. Try putting a mask over the rear aperture of the objective to make it smaller and see if this improves field uniformity.
Just an idea,
Mark Cannell
Michael Hansen wrote:
> Dear Michael.
> We have for our Leica TCS-4d shading-problems that are quite similar to yours. It would be interesting to know if this degree of shading is found on confocal microscopes in general. I will send you our images when you return.
>
> Although we perform off-line shading correction on our images, it would obviously be an advantage if the microscope could be adjusted to minimize shading in the raw images. If you (or someone else) know how this could be done I would appreciate to get the information. Our images are used in ratio imaging which, however, does not eliminate shading. Isn't this to be expected?
>
> Regards,
>
> Michael Hansen
>
> ========================================
> Michael Hansen
> Section of Genetics and Microbiology
> Institute of Ecology
> The Royal Veterinary and Agricultural University
> Thorvaldsensvej 40
> 1871 Frederiksberg C (Copenhagen)
> Denmark
>
> e-mail: [log in to unmask]
> Phone: +45 35 28 26 31
> ========================================
>
> >>> [log in to unmask] 06-12-00 21:01 >>>
> We would love to compare the field uniformity problems we have seen in the
> BioRad Radiance 2000 and the Noran OZ to other confocal microscopes.
>
> If anybody could provide analysis of or images of uniform fields from their
> confocals, we would be happy to do the analysis.
>
> I will be out of the lab from Dec. 8 through 14, but could accept ftp
> delivered images before I leave or after I return.
>
> An example of what we're looking at is at
> http://www.aecom.yu.edu/aif/instructions/BioRad_instructions/installation/un
> iformitydec00/02.htm
>
> BTW, at high zooms, this problem goes away.
>
> Thanks.
> ____________________________________________________________________________
> Michael Cammer Analytical Imaging Facility Albert Einstein Coll. of Med.
> Jack & Pearl Resnick Campus 1300 Morris Park Ave. Bronx, NY 10461
> (718) 430-2890 Fax: 430-8996 URL: http://www.aecom.yu.edu/aif/
|
|
|