Greetings Carl,
For a nice discussion/short review of factors concerning microscope 
objective design and optical resolution you may wish to read the recent 
article in the October issue of Biotechniques:

Abramowitz et al. (2002). BioImaging: Basic Principles of Microscope 
Objectives. Biotechniques, 33,4,772-781.

The rest of this post is a little long winded...

To discuss the resolution of a point scanning microscope we need to 
distinguish between lateral (xy) and axial (xz) resolution.  The 
optical resolution is dependant on the numerical aperture of the 
objective, the wavelength of light being used, any index mismatches in 
the system, and for a confocal scan head, the pinhole among other 
factors related to the scan head optics.  I'll outline some simplified 
rules of thumb.
         Frequently we can make some assumptions (one being that your 
system is perfectly index matched) to easily determine the lateral 
resolution of an objective in the context of a confocal scanning 
system.  The formula below is a simplified calculation Leica provides 
to model the performance of objectives on the SP-2 platform:

                 lateral resolution = 0.4 (wavelength) / numerical 
aperture

         There are different formulas for calculating lateral 
resolution, some more sophisticated than others, but this is the one 
Leica outlines in their documentation.  Most calculations take the form 
of ((some constant)(lambda or wavelength))/2(numerical aperture).  The 
next thing we need to consider is the sampling frequency, in other 
words, even with a diffraction limited spot size if you are only 
sampling a couple pixels in a large scan raster area then you can only 
resolve so much.  According to the Nyquist theorem, the smallest 
optically resolvable distance can be mapped without loss of information 
if it is scanned with about 2 to 3 raster points.  This means your scan 
resolution has to be high enough that you have at least 50% pixel 
overlap; the amount of pixel overlap is dependant on the spot size 
(determined above), the zoom (determines the physical dimensions of the 
scanning area) and the scan resolution (determines the number of 
sampling points/line and number of lines/scan).  Using the information 
above, we assume that the 10x (NA=0.4) objective using the 488nm 
wavelength has a spot size  (lateral optical resolution) of about 488nm 
in an ideal system (specimen located beneath at least .17mm of material 
of a refractive index of 1.514).  The scan field size at zoom =1 is 
1500um and the maximal scan resolution is 2048 pixels per line. 2048 
pixels * .488um spot size = 999um. Clearly this is an example of a 
situation where you are limited by the sampling frequency: 1500um - 
999um = 501um.  501um/2048= .245um between pixels, so your effective 
spot size is .733um.  If you need 2 pixels to resolve one unit, your 
lateral resolution is .733 * 2 or 1.5um.  For maximal resolution, you 
would need to zoom in until (scan resolution) * spotsize (or lateral 
optical resolution) = 2 * (scan field size).  If your scan field size 
is 1500um at 10x * (zoom =1) then the scan field size is 750um at (zoom 
= 2) etc...  Scan field size also decreases proportionally to an 
increase in optical magnification.

         Describing axial resolution is a little more hairy and people 
still argue about it.  The optical axial resolution is limited by the 
sampling freqency as above, except in the z-axis.  The most widely 
accepted formula for estimating axial resolution of an objective in a 
confocal system (as far as I am presently aware) predicts the FWHM of a 
point confocal microscope in reflected mode for a perfect plane mirror 
and is given as 0.45*lambda/(n*(1-cos(alpha)) where n = refractive 
index, lambda = wavelength and NA = n*sin(alpha).  I believe the 
primary reference for this formula is R.Ho, Z. Shao (1991). "Axial 
resolution of confocal microscopes revisited," Optik 88, 4, 147-154.  
If a system isn't index matched, and most thick-section type biological 
specimens for confocal microscopy are not, this estimation will be 
significantly better than the real situation.  For this reason it is 
useful to distinguish between modeling axial resolution and determining 
it quantitatively.  If the refractive index of the sample can be 
approximated, the axial resolution can be determined by imaging 
sub-resolution fluorescent spheres in an xz line scan.
        The scan step increment remains important.  For instance, we can 
pretend a 10x objective (NA 0.4) has an axial resolution of around 5um. 
  So in an axial scan or a 3-D stack of images, your z-galvo step size 
has to be 2.5um or less to realize maximum resolution with the 10x.  
For the sake of comparison, axial resolution of a 63x oil objective in 
an index matched system has been measured at about 350nm using 488nm 
light on our confocal.

Some pertinent references include:

Hell et al., (1993).  Aberrations in confocal fluorescence microscopy 
induced by mismatches in refractive index.  Journal of Microscopy, 
169,3,391-405.

Booth and Wilson. (2001). Refractive-index-mismatch induced aberrations 
in single-photon and two-photon microscopy and the use of aberration 
correction. J. Biomed. Opt. 6,3,266-272.

Hiraoka, Y et al. (1990). Determination of three-dimensional imaging 
properties of a light microscope system. Biophys. J. 57 325-333.

In a conventional confocal or multiphoton rig, you will not be able to 
spatially resolve particles smaller than about 200nm in the lateral and 
axial dimensions unless the spacing is pretty far apart.  Particles 
below the resolution limit will appear as spheres the size of the 
diffraction limited spot size, so you will still be able to detect 
them, you just will not be able to resolve individual particles from 
groups of particles.

Workers are developing clever means to break the resolution limit of 
confocal and multiphoton microscopes.  The smallest excitation spot 
volume I am aware of was a sperical spot of 90-110 nm (.67 attoliters): 
Klar et al., (2000).  Fluorescence microscopy with diffraction 
resolution barrier broken by stimulated emission. PNAS, 
97,15,8206-8210.  There are also the near-field techniques to consider 
such as NSOM, and total internal reflectance/evanescent wave 
stimulation for thin films.

On Monday, Dec 2, 2002, at 11:21 US/Central, Carl Boswell wrote:

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Hi Andrea,
> This discussion is something I've not found a definitive answer for 
> yet.
>> From Inoué, S. (1986) Video Microscopy. Plenum Press, New York, 
>> p.114, the
> lateral resolution is calculated using x=(1.22 lamda)/2NA.  If your 
> numbers
> are plugged into this, it yields 225nm.  I'm particularly concerned 
> about
> this because I have a web page on this topic
> http://www.mcb.arizona.edu/IPC/zoom_Page.htm, and I'm still not 
> comfortable
> with what users might take as the gospel.  Another burning question is
> z-resolution, and how the pinhole size effects this, if at all.  
> Finally,
> why is there uncertanty about how to determine these values?
> Regards,
> Carl
>
> Carl A. Boswell
> Dept. of Mol. Cell. Biology
> Univ. of Arizona
> (520) 626-8469
> [log in to unmask]
>
> ----- Original Message -----
> From: "andrea manazza" <[log in to unmask]>
> To: <[log in to unmask]>
> Sent: Monday, December 02, 2002 8:39 AM
> Subject: Re: confocal resolution
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> Dear Kathy,
> the maximum resolution for optic microscopy should be 200 nm, according
> to standard physics laws: that means that any higher resolution will
> give you no more information on the sample.
> As for useful formulas:
> -       the Nyquist theorem for lateral resolution should be similar to
> X= 0,4*lambda/NA, where NA is numerical aperture of your objective and
> lambda the wavelength. It means that, if you’re working with 63x
> objective and 1,32 as NA, at 488nm, the lateral resolution is 148nm;
> dividing this number by 3, you obtain almost 49 nm. 49 is the raster
> distance you should set in your scan, in order to avoid loss of info
> from the specimen —> it does NOT mean that you will obtain resolution
> higher than 200nm, but only that you’ll obtain good data by scanning
> thrice the 148nm distance obtained before; in order to modify the
> raster distance you should change the electronic zoom  and the scan
> format.
> At least, this is what I've found; if anyone else has nice data, I'll
> be  interested as you in reading more :)
> Thank you and have a nice time.
> AM
>
>
>
>
> dott. Andrea Manazza
> Dipartimento di Oncologia, Università di Torino
> CeRMS, ospedale Molinette
> Via Santena, 7 - 10126 Torino, Italia
> tel: +39-11-6706522
> lab: +39-11-6336859
> fax: +39-11-6336887
>
> ----- Original Message -----
> From: Kathy Spencer <[log in to unmask]>
> Date: Wednesday, November 27, 2002 9:32 pm
> Subject: confocal resolution
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Hello All!
>>         Basic generic question....what is the limit of resolution
>> on a
>> confocal? I'm using an Olympus Fluoview confocal, 60x 1.4NA
>> objective, oil
>> (1.516), pinhole automatically set for 1 Airy disk. For 488nm
>> laser, I
>> calculate 174nm using resolution=wavelength/2NA. But can the confocal
>> resolve more than that? What is a good formula?
>>         Thanks!
>>         Kathy
>>
>>
>>
>>
>>
>>
>>
>>
>> Kathy Spencer
>> The Scripps Research Institute
>> 10550 N. Torrey Pines Road
>> ICND 202
>> La Jolla, CA 92037
>> 858-784-8437
>>
>