Just a tip for the unsaved information on the LEICA:

If you go to the directory /TEMP/DATA/ you can find your pictures saved
as row data. Just be sure you don't run the software before doing that.

Cheers,
Nuno Moreno





-----Original Message-----
From: Confocal Microscopy List [mailto:[log in to unmask]] On
Behalf Of Guy Cox
Sent: quinta-feira, 12 de Junho de 2003 12:23
To: [log in to unmask]
Subject: Re: UPS line stabilizers for Leica TCS SP2


Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal

A Leica TCS system 'typically' has a 4 line argon (which is a powerful
beast) along with green and red HeNe though of course there are any
number of other options.  The question is what are you trying to do with
your system once the power goes out?  Do you really want an extra hour
of full operation?  If you had a UPS which would run just the confocal,
the computer and the cooling fan for the argon laser you would be fine
for a relaxed shutdown with no damage to equipment or loss of data.  And
it would be a MUCH smaller UPS than one to run the lasers.

(For those who are unfamiliar with the Leica, it is possible
to have a LOT of unsaved data in the system which will be lost in a
power outage.  Of course you should save as you go but ....)

Guy Cox




Quoting Robert Atkinson <[log in to unmask]>:

> Hi,
> What lasers are you using? HeNe & HeCd should not be a problem, a few
> 10's of watts each but an Argon Ion will require over 1KW. This means
> a much larger UPS especially for 1 hours operation. Each unit with a
> power cord should have an electrical rating plate, add up all the
> watts  (or VA, KVA) allow 30% for peak demands. This will give you a
> starting point for the size you need. As suggested by Tom, APC's range

> of UPS's are very good. Avoid the cheap units sold at PC outlets,
> these are only meant to support a PC long enough to shut down properly

> without loss of data. Also check with your facilities staff to see if
> there is a generator backed up supply available on site, if so you
> just need a short time UPS to keep things going until the generator
> kicks in.
>
> Regards,
> Robert.
>
>
> -----Original Message-----
> From: Donnelly, Tom [mailto:[log in to unmask]]
> Sent: 11 June 2003 19:23
> To: [log in to unmask]
> Subject: Re: UPS line stabilizers for Leica TCS SP2
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> Try http://www.apc.com/index.cfm  they have a full line of stabilizers

> UPS systems.  Leica and the laser mfg.  should be able to provide you
> with the power load you will need to support.
>
> Tom
>
> -----Original Message-----
> From: Hollie Donze [mailto:[log in to unmask]]
> Sent: Wednesday, June 11, 2003 10:37 AM
> To: [log in to unmask]
> Subject: UPS line stabilizers for Leica TCS SP2
>
>
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I was wondering if anyone could recommend manufacturers of UPS systems

> which will stabilize (2) 220V lines to a Leica TCS SP2 system with 3
lasers and
> provide battery backup for at least 30 minutes- 1 hr?    I've already
> contacted Leica and they recommend getting recommendations from users!

> Thank you in advance!
>
>
>
> Hollie Hale-Donze, Ph.D.
>
> Socolofsky Microscopy Center
>
> 202 Life Sciences Building
>
> LSU
>
> Any opinions expressed in this email are those of the individual and
> not necessarily Genetix Ltd (Genetix) or any company associated with
> it. This email and any files transmitted with it are confidential and
> solely for the use of the intended recipient. If you are not the
> intended recipient or the person responsible for delivering to the
> intended recipient, be advised that you have received this email in
> error and that any use is strictly prohibited. If you have received
> this email in error please notify Genetix by telephone on +44 (0)1425
> 624600.
>
>
> The unauthorised use, disclosure, copying or alteration of this
> message is strictly forbidden. This mail and any attachments have been

> scanned for viruses prior to leaving the Genetix network. Genetix will

> not be liable for direct, special, indirect or consequential damages
> as a result of any virus being passed on, or arising from alteration
> of the contents of this message by a third party.
>
>


--
Associate Professor Guy Cox
Electron Microscope Unit, F09
University of Sydney NSW 2006
+61 2 9351 3176

Until 25th July:
Instituto Gulbenkian de Ciencia, PT-2780-156, Oeiras, Portugal.
+351 21 446 4638 (office) or +351 91 401 5726 (mobile)
Fax: +351 21 440 7970

ATOM RSS1 RSS2