You can always expect a deviation at the high and low ends since there
will be, statistically, some pixels over 255 and under 0 which will
therefore be truncated. But I must admit I'm surprised it reaches
down as far as 200 - playing about on an old Bio-Rad I found that
things seemed pretty linear as long as I kept the maximum below ~230.
On a modern 12-bit system you should anyhow have a pretty reasonable
linear range to play with.
Guy
Quoting Greg Martin <[log in to unmask]>:
>> One more note on linearity: We've found that on our Leica SP1 (8-bit
> system) that the response of the PMTs is only linear between 20 and 200 grey
> levels (these limits depend on how fussy one is about curve-fitting). Same
> results whether we use a graded concentration of dye in solution, or adjust
> excitation with the AOTF (assuming AOTF attenuation is linear). Does anyone
> know if this is typical?
>
> Peace! Greg.
--