>Hello all!
>
>I have used the latex beads inside the agarose
>gel and I have measured the penetration depth of
>NA: 1.2 60X lens about 600um,could it be wrong?
>
>Thanks
>Sarah
Dear Sara,
You don't say, but I believe that the (Nikon?)
objective you describe is supposed to be used
with a coverslip.
If you just use it just with agarose, two things
will happen: You will have VERY high spherical
aberration (and therefore z-direction
measurements are not worth much) and, because the
coverslip is missing, the working distance will
appear to be more than the (about ) 200 µm that
the lens was designed for.
Add a 170 µm coverslip on top of the agarose, and
you should be able to see about 200 µm below the
agarose/coverslip interface.
Cheers,
Jim P.
--
**********************************************