Kate,
the objective kills quite some light, its transmission at 800nm is about
60% according to the Zeiss website. Check the database and your place
for objectives with better IR transmission, i.e. the Neofluar 40x/1.3
(80% at 800nm). This might already help to boost the detection a bit.
Michael
Kate Phelps wrote:
> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
> I neglected to mention that we can take very nice pictures of
> immunofluorescence labeling with IRDye 800 goat anti-rabbit secondary
> antibody with this set up. So it is a problem of optimizing to detect a very
> weak signal, not whether we can do it at all.
> Thanks for everyone's help. Please keep the suggestions coming!
>
> Kate