Dear Aryeh,
We regularly use a 20x LWD on our IX81 without any problem. You should
be aware that the working distance of a LWD objective is several mm, so
you need to be considerably further away from your sample compared to
high-NA imaging. Eventually you simply need to lower your objective
(using the focus wheel). if you can't lower your objective enough (to
have a proper image) then eventually a focus-limit has been
set/activated on your microscope.
Cheers Gabor
> I am a problem imaging with the Olympus LWD (long working distance)
> objectives on an IX81 frame. I have image well with the high NA
> objectives (oil and air),
> which are used when imaging through thin coverslips. However, with the
> LWD objectives, I cannot raise the turret enough to get an image.
> This is puzzling, because I thought that the objectives were designed
> to be parfocal, so that if I have focus for one objective, I will have
> focus for the other at the same turret position. Yet, if I put a
> regular slide with a coverlip on the scope, I can image it with the
> 20x close focus air objective, but then when I switch to the LWD
> objective ( and set the cover thickness ring appropriately for a
> coverslip, I cannot even get close to an in-focus image.
>
> The stage is a LUDL motorized stage, and it may be higher than the
> regular mechanical stage, but this does not explain why the close
> focus high NA objectives can focus, and the LWD objective dont.
>
> Thanks in advance for putting me straight on what is probably
> something obvious.
>
> --aryeh
--