Since I'm reading this very chapter now,Here are the possibly useful things I extracted from it.


Microspores and Pollen Grains

1. Pollen grains are usually highly absorbing, scattering and pigmented(hence strongly autofluorescenting in visible spectrum).If staining is necessary, considering NIR dyes.
2. Pollen grains/microspores can be handled the same way as suspension-cultured cells.
3. If possible, fix,stain and clear tissue with methyl-salicylate, so as to match the optical property as closely as possible to the design criteria of lens.
4. Multi-photon imaging and adaptive deconvolution might be helpful for obtaining high resolution image


(Handbook of Biological Confocal Microscopy,3rd Edition, Chapter21
Interaction of Light with Botanical Specimens, Chapter 21. P431.)



Edna


--------------------
Hi Doug,
I have been doing some 3D pollen imaging with good success. So far I imaged 
pollen grains up to 50 um in diameter, and it seems I get almost the same 
resolution on the far side as I get on the side closest to the coverslip.

The sample prep was the key. 

1st, the pollen I received from our pollen expert was extracted (not sure of 
the exact protocol, apparently somehting the pollen researchers do routinely), 
and stored in glacial acetic acid. The result is that the highly absorbent and 
scattering contents of the pollen grain has been removed, and all that is left is 
the pollen "shell" with its characteristic shape and features. It has enough 
autofluorescence for confocal imaging without any staining.

The pollen is gradually re-hydrated and infiltrated with 2-2-thiodiethanol 
mounting medium.
below is the prep protocol: The microwave just accelerates the process, you 
could just do longer time each step without the MW.

100 µl of the pollen suspension was placed in a microcentrifuge tube and 
gradually re-hydrated by adding increasing amounts of water (from 10 µl to 
600 µl), pulse-mixing (vortexing) and irradiating in a Pelco Biowave (Ted Pella 
Inc., Redding, CA) laboratory microwave processor for 1 min at 230W to 
accelerate the diffusion process. Total 12 steps were performed. Pollen was 
then spun down by centrifugation at 500 x g for 1 min, resuspended in 
phosphate buffered saline (PBS; 140 mM NaCl, 3 mM KCl, 10 mM Na2HPO4, 2 
mM KH2PO4) and microwaved as above.

Subsequently, the pollen was gradually infiltrated with 2,2’-thiodiethanol 
(TDE), a mounting medium for high-resolution microscopy (Staudt et al., 
2007). The pollen was spun down as above before each step. The pollen pellet 
was resuspended using 10%, 25%, 50% v/v TDE/PBS mixture, and finally three 
times in 97% v/v TDE/PBS, In each step, 1 min microwave irradiation at 230W 
was used after resuspension.  
The pollen was then spun down and the pellet resuspended and stored in 97% 
TDE/PBS.

Confocal Microscopy (Inverted microscope)
For microscopy, a droplet of the pollen suspension was applied to a coverslip-
bottom imaging chamber (coverslip thickness ~175 µm) and the pollen 
immobilized by placing a small coverslip on the top of the droplet, with a small 
dab of dried nail polish on one side as a spacer so that the weight of the 
coverslip does not cause flattening of the pollen grain. Close to the spacer, 
there layer of the mounting medium is too deep, and the pollen is floating 
freely and cannot be imaged. On the other end of the coverslip, there is no 
spacer and the pollen is squished, deformed. So I try to find a position in 
between, where the pollen is not moving, but is not visibly squished.

Imaging was done using Olympus FV1000 laser scanning confocal microscope 
system attached to an Olympus IX81 inverted microscope with a 100x/1.4 oil 
immersion objective (UPLSAPO 100x/1.4). Fluorescence was excited using 488 
nm Argon ion laser, the fluorescence detector was set to 500-600nm 
bandpass. Laser output was programmed  to increase as we go deeper from 
the surface in order to compensate for the loss of signal. Typically, the laser 
output would double from 8 to 16% for a Z-stack going from 0 to 50 um depth.
 
Confocal scanning was performed in the photon counting mode, scan speed 
(pixel dwell time) was 20 µs/pixel or 10 µs/pixel. The confocal aperture was 
set to 120 µm, which corresponds to 0.75 Airy Unit. The confocal zoom and 
scan size was set to achieve 65nm pixel size in XY. The z-step was set to 
130nm. 
Typical acquisition time was between 20 and 50 min per stack, depending on 
the size of the pollen grain and scan speed (i.e., signal intensity).

You may not need to do such a fine XYZ step, but I have to say the resulting 
datasets look great;

Sincerely,

Stan Vitha
Microscopy and Imaging Center, 
Texas A&M University


On Tue, 6 Jan 2009 10:47:06 -0700, [log in to unmask] wrote:

>I have very close to zero experience prepping plant tissues for any type of
>microscopy.  I've been contacted by a botanist that is interested in looking
>at a variety of pollen types and he's never done confocal.  He'd like to
>look at the 3D structures and it sounds like good 3D may very well be
>important.
>
>
>
>I know from limited experience that the far side of a pollen grain often
>doesn't look as good in a 3D rendering, presumably due to spherical
>aberration, scattering, and other optical effects.  I gather from the
>resources I have that pollen is highly autofluorescent across the entire VIS
>spectrum (probably why the vendors love to use it for demos).
>
>
>
>Any tips, suggested review articles, web sites, etc that people could point
>me to for sample prep techniques?  Any microscopy caveats to be aware of?
>
>
>
>Many thanks.
>
>Doug
>
>
>
>^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^^
^^^^^^^^^^^^
>
>Douglas W. Cromey, M.S. - Assistant Scientific Investigator
>
>Dept. of Cell Biology & Anatomy, University of Arizona
>
>1501 N. Campbell Ave, Tucson, AZ  85724-5044 USA
>
>
>
>office:  AHSC 4212         email: [log in to unmask]
>
>voice:  520-626-2824       fax:  520-626-2097
>
>
>
>http://swehsc.pharmacy.arizona.edu/exppath/
>
>Home of: "Microscopy and Imaging Resources on the WWW"
>
>
>
>

ATOM RSS1 RSS2