I just realized, I had nearly forgotten to follow up on this. Yes: it
turned out I had a source of stray light in the dark box surrounding
the scope: the glow in the dark tags used to label the filter wheel
positions on the original Olympus body. Removing them gave me a major
improvement, blackening several polished (and thus reflecting)
surfaces in the detection path did the rest.
Just in case someone stumbles over similar problems.
Thanks for all helpful hints,
Christian
> My thought, however, was more along the lines of stray light. With
> wide-field collection you have to be absolutely sure that room light
> (even from the LEDs on some equipment in an otherwise dark room) is
> not entering the objective or reaching your PMTs through gaps in the
> light guides. Is there some what you can wrap the PMT in black tape
> just for comparison?
|
|
|