And also do not use the computer for anything during scanning. It can give
you artifacts in the final image.
~aleem
On Fri, Dec 14, 2012 at 12:03 PM, Aleem Syed <[log in to unmask]> wrote:
> Increase your exposure time per pixel. It may help, did help in my case.
>
> Good luck.
>
> ~Aleem
>
>
> On Fri, Dec 14, 2012 at 11:59 AM, Craig Brideau <[log in to unmask]>wrote:
>
>> *****
>> To join, leave or search the confocal microscopy listserv, go to:
>> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> *****
>>
>> It is also possible that the electronics are not grounded properly. You
>> can get electronic noise from an improper ground loop. It's also possible
>> you have light leaking in from somewhere; I've gotten stripes in images
>> from flashing LEDs on equipment near the microscope. A quick piece of
>> black tape over the offending LED and no more noise!
>>
>> Craig
>>
>>
>> On Thu, Dec 13, 2012 at 11:46 PM, Pascal Weber <[log in to unmask]
>> >wrote:
>>
>> > *****
>> > To join, leave or search the confocal microscopy listserv, go to:
>> > http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>> > *****
>> >
>> > Hi everebody,
>> >
>> > I had heard about the problem of sensitivity of PMT has been set by a
>> > technician
>> > linearization sensitivity. (Procedure to apply)
>> > For sensitivity to mobile I am very surprised and I do the same test!
>> >
>>
>
>
>
> --
>
>
--