Dear Jeff,

I agree with Guy about the Kohler but there is 
also the matter of wavelength. DIC prisms do work 
with "white" light but they are usually designed 
to work best at a particular wavelength (often 
the green line of the Hg source, 546nm). I don't 
know the emission spectrum of your LED (or how 
hot you have your tungsten-halogen, as this 
affects it spectrum, IR filter? UV filter?) and I 
don't know the response of your detector, but if 
seeing the greatest possible contrast is 
important, you might optimize the wavelength of 
the operation as well as being sure that Kohler 
is working properly

Another thing to keep in mind is that DIC depends 
on keeping two light paths separate by having 
their polarization be at 90 degrees to each 
other. Problems can stem from the fact that the 
degree to which light is reflected when passing 
through a glass-air surface varies with the pol 
axis. This is not a major problem on low-NA 
lenses because the incidence angle is near normal 
and reflection is low (especially on modern 
coated optics). However, on high-NA lenses, rays 
near the edge of the aperture impinge at angles 
where the differential reflection of the two ray 
bundles becomes significant. This can be seen by 
looking at the BFP using the phase telescope 
(Bertrand lens?) with only the polarizer and 
analyzer in place and adjusted for maximum 
extinction (no DIC prisms and a clear glass 
specimen, but an oiled condenser open to the same 
NA as the objective).

If the problem is significant, you will see what 
is called The Maltese Cross, which comes about 
because 4 symmetrically located, blurry, 
more-or-less circular lighter blobs occur in the 
NA regions where particular ray-bundles are 
depleted. Of course, this also happens when you 
are using the same optics with the DIC prisms in 
place.  The result is that the two bundles are no 
longer quite independent and this reduces the 
contrast. Shinya Inoue developed a corrector lens 
that Nikon used to sell to reduce this problem, 
but I don't know if it was ever upgraded to the 
newer scopes (maybe someone from Nikon can tell 
us?). And while we are on the topic of pol, DIC 
is best performed using Pol (strain-free) 
objectives because strain-birefringence can also 
mess up pol performance. I am not clear as to 
whether your 1.45 lens meets this requirement.

Finally, if you interested in seeing individual 
microtubules, you will need more than eyeballs 
and good optics.  MTs are very small and don't 
produce much contrast to start off with. They 
were initially only made visible by virtue of 
video-rate electronic contrast enhancement that 
involved not only using a very bright light 
source (to increase signal levels and reduce the 
relative effect of Poisson Noise in the detector, 
and to reduce this further, one could 
exponentially average the image electronically 
over many video frames. Usually one used a 
mercury source, but if you do this, be sure to 
include a UV filter (and perhaps an 
"interference-green" to isolate the 546nm line) 
so you don't damage your polarizer.). The 
resulting high "average brightness" was then 
electronically subtracted off so that one could 
expand the contrast of the signal that remained.

This procedure invariably produced a very blotchy 
image. Small imperfections in the optics (dust, 
bubbles, scratches) create low contrast and 
generally out-of-focus features referred to as 
mottle. The contrast of mottle is so low that it 
is only visible after the sort of contrast 
enhancement just described. What made DIC so 
suitable for viewing small features by 
video-enhancment was its very narrow depth of 
field. By slightly changing the focus, one could 
form an image in which the ONLY features recorded 
where those due to mottle. This image could then 
be averaged, stored and subtracted from the live 
image (often this difference image was also 
averaged to further reduce noise). The result was 
essentially the image contrast caused only by the 
"now-in-focus" MT, a feature that might have had 
only 1% contrast before enhancement.

Good luck,

Jim Pawley

>Jeff,
>
>	If you want to get decent contrast in DIC 
>you need a high NA condenser with a high NA 
>objective.  Of course the plate must be correct 
>for the objective, and Köhler illumination must 
>be correctly set up.  My guess is that that 
>wasn't true for your LED source, and you weren't 
>filling the BFP.  Don't take this personally but 
>the widespread use of fluorescence has made many 
>microscopists become slack about setting up 
>Köhler illumination! 
>
>				Guy
>
>Guy Cox, Honorary Associate Professor
>School of Medical Sciences
>
>Australian Centre for Microscopy and Microanalysis,
>Madsen, F09, University of Sydney, NSW 2006
>
>-----Original Message-----
>From: Confocal Microscopy List 
>[mailto:[log in to unmask]] On 
>Behalf Of Jeff Spector
>Sent: Saturday, 29 November 2014 1:22 PM
>To: [log in to unmask]
>Subject: DIC condensers
>
>*****
>To join, leave or search the confocal microscopy listserv, go to:
>http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
>Post images on http://www.imgur.com and include the link in your posting.
>*****
>
>Greetings,
>    This isn't exactly a "confocal" questions but 
>I know a lot of "micoscopy gurus" live on this 
>list so I thought it a good place to ask this. I 
>have a colleague who is trying to image 
>individual (i.e. small and diffraction
>limited) microtubules in a flow chamber by using 
>DIC. They are currently using a 100x 1.45 oil 
>Objective, but only a .52 LWD condenser. They 
>were using a solid state light source but 
>couldn't get good image so we switched to a lamp 
>for illumination and the images are much better, 
>and we can now see the microtbules but there 
>still isn't a lot of contrast.  My question is, 
>is it worth it to go to a high NA (perhaps oil 
>immersion) condenser, and can anyone think of 
>why the lamp would give a better DIC image than 
>a solid state light source?
>thanks in advance for the help...
>  -Jeff


-- 
               ****************************************