Hi Ben,
In my experience the critical point is the sample drifting in the first 20 min after being put on the stage. Then the microscope heating up. We have a combined SIM/PALM system where we changed the cameras to water cooling to prevent this. I would expect temperature fluctuations in the room of plus minus 1 degree to be fine and occasionally our air-conditioning does not manage to cope, but the images are still good. Multi-colour beads in the sample help to align the different channels.
Best wishes
Andreas
-----Original Message-----
From: "Benjamin Abrams" <[log in to unmask]>
Sent: 04/02/2015 19:20
To: "[log in to unmask]" <[log in to unmask]>
Subject: SIM environment requirements
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****
Hi All,
We are looking in to getting a SIM system and see that the room requirements
are really strict, +/- 0.5 deg C. This would require significant room
modification (possibly several $100K worth). +/- 1.0 deg C is more doable.
What is the consensus out there? How necessary is +/- 0.5? Can anyone
comment who has experience running their system in a room outside of spec
(even temporarily)?
Thanks,
Ben
|
|
|