Guillaume,
First, you should provide additional information, in particular which
microscope are you using? That aside, the whole point of confocal is that
you can optically section without deconvolution. However, if you use a
large pin hole (iris setting on a BioRad, e.g.) then deconvolution can be a
useful tool allowing you to acquire higher photon counts (hence lower
noise) while still getting good optical sectioning. However, I am of the
opinion that you must calibrate the x-y-z-axes point spread function for
each iris setting and lens that you use so that an accurate deconvolution
can be calculated. This can be done with a sample using small 100 nm
fluorescent beads.