PROBEUSERS Archives

December 2009

PROBEUSERS@LISTS.UMN.EDU

Options: Use Proportional Font
Show Text Part by Default
Show All Mail Headers

Message: [<< First] [< Prev] [Next >] [Last >>]
Topic: [<< First] [< Prev] [Next >] [Last >>]
Author: [<< First] [< Prev] [Next >] [Last >>]

Print Reply
Subject:
From:
Ellery Frahm <[log in to unmask]>
Reply To:
JEOL-Focused Probe Users List <[log in to unmask]>
Date:
Tue, 15 Dec 2009 07:31:29 -0600
Content-Type:
text/plain
Parts/Attachments:
text/plain (37 lines)
JEOL Probe Users Listserver

Moderator: Ellery Frahm, [log in to unmask],
Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html



*

Hi Steve,

Yes, I've seen that before on our 8900.  It is dirt in the lower  
column charging, which deflects the beam as it becomes increasingly  
charged.  Then it discharges, and the beam "snaps back" to its  
starting position again.  You'll probably see it happen faster with a  
larger beam current.  The dirt is most likely in the lower column,  
probably just inside the objective lens pole piece or very near it --  
something from the chamber deposited there.  Cleaning the objective  
lens pole piece should fix that problem.

Best,
Ellery

-----------------
Ellery Frahm
Senior Research Fellow, Department of Geology & Geophysics
Manager & Principal Analyst, Electron Microprobe Lab
University of Minnesota - Twin Cities campus
Lab website: http://probelab.geo.umn.edu/
Personal: http://web.mac.com/elleryfrahm/

ATOM RSS1 RSS2