Hi Jens,
A shame about your iced up EDS detector; very annoying for light
elements and L lines.
I believe you cannot warm up a SiLi EDS detector under power, but I
believe it is OK to warm up a SiLi EDS detector if it is not under
power. I have done it myself a few times with success. My general
procedure is:
To warm up EDS:
(A) Turn OFF the power on the EDS detector control box. I turn off
everything that is connected to the detector, the one that supplies
the HV bias is the most important.
(B) After a suitable pause, I carefully unplug all the leads and
connectors to the EDS detector unit itself. Mine has a multi-pin plug
and a BNC (HV) plug. I use electrical tape/electrical boxes to secure
and insulate all the loose bits and bobs.
(C) Then let the EDS detector warm up slowly. This may take a few days
or more. I put up signs saying "DO NOT FILL EDS DETECTOR - DO NOT
CONNECT TO POWER" etc. Some people blow out/warm up the LN2 from the
detector with air, but this can be dangerous if done too vigorously,
as it can shower the whole lab with LN2. I usually take the lid off
the EDS detector to speed up the warming. I also tie some cloth around
the outside of the dewar to catch the condensate. I have always left
my detector connected to the SEM and under vacuum.
(D) When the detector is warm, then make sure it is dry inside the
dewar. Ice inside the dewar causes noise. The ice on the detector
should now be gone. I clean out any solid debris from the dry dewar.
To cool down EDS:
(E) Fill the EDS detector dewar with LN2, and let it cool down for at
least two hours. Top up the LN2 after one hour just to make sure it
has enough LN2.
(F) After you are sure the EDS is cold, and have double checked all
the power sources are still off, then reconnect all the plugs and
leads to the EDS detector.
(G) Turn on power to the EDS control box.
(H) Check and calibrate. I use copper (a Cu TEM grid) at 20kV with a
20-30% dead time. In my system the Cu L line should be more than twice
the Cu K line.
A few words of caution:
Please use all care and common sense. Remember I have not seen your
system. Some electrical components may still hold a charge after
disconnecting. There may be local electrical licencing issues. Be
aware of the safety issues with liquid nitrogen, cryo-burns and the
displacement of oxygen.
Finally, when I worked out my plan of attack, I described the detail
of what I wanted to do and sent it to the local EDS agent. They read
it and offered their opinion and approval before I proceeded.
I hope this helps - at least as a starting point.
cheers,
Ron
Quoting "Jens C. Andersen" <[log in to unmask]>:
> JEOL Probe Users Listserver
>
> Moderator: Ellery Frahm, [log in to unmask],
> Electron Microprobe Lab, University of Minnesota
>
> Post a message: send your message to [log in to unmask]
>
> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
>
> On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html
>
>
>
> *
>
> Dear colleagues.
>
> I'm looking for information on how to safely de-ice an EDS detector.
> The detector is of the lithium drifted silicon detector type that
> cannot be brought up to room temperature.
>
> Any advice is appreciated.
>
> Thanks,
>
> Jens
>
>
> Jens C. Andersen
> Camborne School of Mines
> School of Geography, Archaeology, and Earth Resources
> University of Exeter
> Cornwall Campus
> Penryn, Cornwall
> TR10 9EZ
> United Kingdom
> Tel. +44 (0)1326 371 836
> Fax. +44 (0)1326 371 859
>
> http://www.exeter.ac.uk/cornwall/csm
>
> Visit the virtual Skaergaard intrusion at http://www.skaergaard.org
|