PROBEUSERS Archives

July 2011

PROBEUSERS@LISTS.UMN.EDU

Options: Use Monospaced Font
Show HTML Part by Default
Condense Mail Headers

Message: [<< First] [< Prev] [Next >] [Last >>]
Topic: [<< First] [< Prev] [Next >] [Last >>]
Author: [<< First] [< Prev] [Next >] [Last >>]

Print Reply
MIME-Version:
1.0
Sender:
JEOL-Focused Probe Users List <[log in to unmask]>
Subject:
From:
"Blake, Kevin" <[log in to unmask]>
Date:
Tue, 5 Jul 2011 20:28:26 -0700
In-Reply-To:
Content-Type:
multipart/alternative; boundary="_000_CA3A137A4AB7kevinblakejcueduau_"
Reply-To:
JEOL-Focused Probe Users List <[log in to unmask]>
Parts/Attachments:
text/plain (3962 bytes) , text/html (5 kB)
JEOL Probe Users Listserver

Moderator: Ellery Frahm, [log in to unmask],
Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html



*

Hi Matthew,
Shutdown ca. 30  minutes from start up may be more to do with the system thinking the diff’ pumps are not up to temperature. Have you checked/replaced the thermocouples or tricked them as per Ritchies’ suggestion?
Cheers
Kevin


On 6/07/11 2:58 AM, "Matthew Beck" <[log in to unmask]> wrote:

JEOL Probe Users Listserver

Moderator: Ellery Frahm, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html

* Hi Ritchie,

I am Anthony's partner in getting the probe functional. The model is a JXA-733. We have tried running with both the DP on and off. The DP does get hot when turned on but the system still shuts down after 30 minutes of continuous operation. We wish to pump continuously on the system at the advice of our supervisor. The system originally failed when the compressor failed allowing the the RP to pump on atmosphere for days on end. This is in turn caused the RP to fail. It has since been replaced. It is his belief that the chambers are now inundated with oil vapor and must be purged through continuous RP pumping.

Best,

Matthew Beck
Illinois Institute of Technology

On Tue, Jul 5, 2011 at 1:26 AM, Ritchie Sims <[log in to unmask]> wrote:
JEOL Probe Users Listserver

Moderator: Ellery Frahm, [log in to unmask],
Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html



*

Are you running it with the DPs switched off and cold?

My 840 switches off if the DPs are not warmed up within 30 minutes, but I can leave it pumping the cold DPs (this is all using manual control of the valves, of course) overnight simply by jumping the warmup (not the overtemp) temperature sensing switches on the DPs.

Maybe your superprobe (which model?) is the same.


BTW, I'm not sure that it's such a great idea to suck on the cold DPs for extended periods as I suspect the DP fluid will get somewhat contaminated with RP oil by backstreaming. But I don't have any foreline traps, maybe you do.


cheers
Ritchie Sims
University of Auckland
________________________________________
From: JEOL-Focused Probe Users List [[log in to unmask]] on behalf of Anthony J. Ruth [[log in to unmask]]
Sent: 02 July 2011 03:49
To: [log in to unmask]
Subject: [PROBE-USERS] disabling 30 minute timer

JEOL Probe Users Listserver

Moderator: Ellery Frahm, [log in to unmask],
Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html



*

The superprobe I am working on is having vacuum trouble. We need to run the
roughing pump for an extended period of time to reach low vacuum and turn on the
diffusion pump, but there is a timer that turns off the system if high vacuum is
not reached in 30 minutes. How can I disable the timer? I have tried adjusting
the potentiometers on the manual pump control panel so that all three pressure
meters report high vacuum, but it still did not disable the timer.




********************************
Dr Kevin Blake
Director
Advanced Analytical Centre
James Cook University
Townsville
Queensland 4811
07 4781 4864
[log in to unmask]

Linked Laboratory to:
Australian Microscopy and Microanalysis Research Facility
AMMRF
ammrf.org.au <http://www.ammrf.org.au/>
********************************



ATOM RSS1 RSS2