OK, here are the photos.
----------------------------------------------------------------
Ioannis Baziotis [log in to unmask]
Assistant Professor of Mineralogy and Petrology
Agricultural University of Athens
Laboratory of Mineralogy and Geology
Athens, 11855 Greece
(+30)210-529-4155
Twitter: https://twitter.com/EGU_GMPV
Personal site: http://ibaziotis.wix.com/ioannis-baziotis
----------------------------------------------------------------
*Next dream is "on-the-go" .........**mid -April 2019*
Στις Πέμ, 3 Ιαν 2019 στις 6:42 μ.μ., ο/η Ioannis Baziotis <
[log in to unmask]> έγραψε:
> Here is a set of consecutive photos (starting from low mag X40). As you
> see, part of the image (at left) is dark (the same occurs both in SE and
> BSE). Also, toward higher mag, the image becomes "blurry" (or something
> like that).
> ----------------------------------------------------------------
> Ioannis Baziotis [log in to unmask]
> Assistant Professor of Mineralogy and Petrology
> Agricultural University of Athens
> Laboratory of Mineralogy and Geology
> Athens, 11855 Greece
> (+30)210-529-4155
> Twitter: https://twitter.com/EGU_GMPV
> Personal site: http://ibaziotis.wix.com/ioannis-baziotis
> ----------------------------------------------------------------
> *Next dream is "on-the-go" .........**mid -April 2019*
>
>
>
> Στις Πέμ, 3 Ιαν 2019 στις 6:35 μ.μ., ο/η William J Mushock <
> [log in to unmask]> έγραψε:
>
>> Could you provide a picture or video of what you describe as internal
>> reflections? And, are you convinced this is the result
>> of the same issue? Confession: In a former life I was a field service rep
>> for JEOL and had a lot of experience on the 8900s.
>> I think it would be a big help to see what you are describing.
>>
>> Bill
>>
>> On Thu, Jan 3, 2019 at 10:01 AM Collins, Keith W. <
>> [log in to unmask]> wrote:
>>
>>> Unfortunately, It may be time to schedule a service visit from JEOL.
>>>
>>>
>>>
>>> *From:* JEOL-Focused Probe Users List <[log in to unmask]> *On
>>> Behalf Of *Ioannis Baziotis
>>> *Sent:* Thursday, January 3, 2019 3:55 AM
>>> *To:* [log in to unmask]
>>> *Subject:* [EXTERNAL] Re: [PROBEUSERS] Re: [PROBEUSERS] Re:
>>> [PROBEUSERS] URGENT HELP: JEOL JXA 8900 Superprobe
>>>
>>>
>>>
>>> We have cleaned and reseat all pc boards. Checked all the fuses at the
>>> front and back of the control unit but nothing changed. Any other ideas to
>>> test?
>>>
>>>
>>>
>>> Στις Παρ, 28 Δεκ 2018, 21:30 ο χρήστης Collins, Keith W. <
>>> [log in to unmask]> έγραψε:
>>>
>>> You can generally visually look at the fuse and see if the wire has
>>> burned out (magnifying glass helps) or use a volt ohm meter to test,. IF
>>> the fuse is bad you replace it with the same rated fuse. Fuses may be
>>> located in a block of fuses, on the circuit board or in a fuse holder in a
>>> wire (usually a twist to open). Fuses should be marked on the wiring
>>> diagram. I would suggest a service call.
>>>
>>>
>>>
>>> *From:* JEOL-Focused Probe Users List <[log in to unmask]> *On
>>> Behalf Of *Ioannis Baziotis
>>> *Sent:* Friday, December 28, 2018 10:56 AM
>>> *To:* [log in to unmask]
>>> *Subject:* [EXTERNAL] Re: [PROBEUSERS] Re: [PROBEUSERS] URGENT HELP:
>>> JEOL JXA 8900 Superprobe
>>>
>>>
>>>
>>> Ok.
>>>
>>> 1) how do you check them?
>>>
>>> 2) how do you replace them?
>>>
>>>
>>>
>>> Στις Παρ, 28 Δεκ 2018, 20:37 ο χρήστης Collins, Keith W. <
>>> [log in to unmask]> έγραψε:
>>>
>>> Also check fuses a blown fuse to the scan coils can result in half an
>>> image.
>>>
>>>
>>>
>>> *From:* JEOL-Focused Probe Users List <[log in to unmask]> *On
>>> Behalf Of *Ioannis Baziotis
>>> *Sent:* Friday, December 28, 2018 9:22 AM
>>> *To:* [log in to unmask]
>>> *Subject:* [EXTERNAL] Re: [PROBEUSERS] URGENT HELP: JEOL JXA 8900
>>> Superprobe
>>>
>>>
>>>
>>> Great...i`ll try it. so, which PC boards you exactly mean? The one
>>> seeting at the front part of the control unit, or at the back, too?
>>>
>>> ----------------------------------------------------------------
>>> Ioannis Baziotis [log in to unmask]
>>> Assistant Professor of Mineralogy and Petrology
>>> Agricultural University of Athens
>>> Laboratory of Mineralogy and Geology
>>> Athens, 11855 Greece
>>> (+30)210-529-4155
>>> Twitter: https://twitter.com/EGU_GMPV
>>> Personal site: http://ibaziotis.wix.com/ioannis-baziotis
>>>
>>> ----------------------------------------------------------------
>>>
>>> *Next dream is "on-the-go" .........mid -April 2019*
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>> [image:
>>> https://ipmcdn.avast.com/images/icons/icon-envelope-tick-round-orange-animated-no-repeat-v1.gif]
>>> <https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail>
>>>
>>> Απαλλαγμένο από ιούς. www.avast.com
>>> <https://www.avast.com/sig-email?utm_medium=email&utm_source=link&utm_campaign=sig-email&utm_content=webmail>
>>>
>>>
>>>
>>> Στις Παρ, 28 Δεκ 2018 στις 5:56 μ.μ., ο/η William J Mushock <
>>> [log in to unmask]> έγραψε:
>>>
>>> Ioannis,
>>>
>>>
>>>
>>> Turn off console power, remove the lower front cover and try re-seating
>>> the PC boards. Sounds like you may have developed
>>>
>>> a bad contact (oxidation) on the edge connector of one of the boards.
>>>
>>>
>>>
>>>
>>>
>>> On Fri, Dec 28, 2018 at 8:25 AM Ioannis Baziotis <[log in to unmask]>
>>> wrote:
>>>
>>> Dear Colleagues,
>>>
>>> I have noticed the following problem on my JEOL JXA-8900 Superprobe.
>>> At the lowest magnification (40X), the 30% of the screen (left part) is
>>> black (either under SE or BSE). Please note, that under OM (reflected
>>> light) the image looks perfect. Also, at higher mag the horizontal movement
>>> (when inspecting a sample) is like having internal reflections. The
>>> filament has been replaced very recently, however, i had the same problem
>>> still occurring since the last week with the previous filament. Also, at
>>> higher mag (from 950X to 1000X and 9500X to 10000X) there is a considerable
>>> shift on horizontal axis. I tried the objective lense aperture (X and Y)
>>> however, nothing changed. Any idea for the problem and how can i fix it?
>>>
>>> All the best.
>>>
>>> ****
>>> JEOL Probe Users Listserver
>>>
>>> Moderator: Anette von der Handt, [log in to unmask],
>>> Electron Microprobe Lab, University of Minnesota
>>>
>>> Post a message: send your message to [log in to unmask]
>>>
>>> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
>>>
>>> On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html
>>>
>>> *
>>>
>>> **** JEOL Probe Users Listserver
>>>
>>> Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe
>>> Lab, University of Minnesota
>>>
>>> Post a message: send your message to [log in to unmask]
>>>
>>> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
>>>
>>> On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html
>>>
>>> *
>>>
>>> **** JEOL Probe Users Listserver
>>>
>>> Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe
>>> Lab, University of Minnesota
>>>
>>> Post a message: send your message to [log in to unmask]
>>>
>>> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
>>>
>>> On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html
>>>
>>> *
>>>
>>>
>>> ********************************************************************
>>> This message does not originate from a known Department of Energy email
>>> system.
>>> Use caution if this message contains attachments, links or requests for
>>> information.
>>>
>>> ********************************************************************
>>>
>>> **** JEOL Probe Users Listserver
>>>
>>> Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe
>>> Lab, University of Minnesota
>>>
>>> Post a message: send your message to [log in to unmask]
>>>
>>> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
>>>
>>> On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html
>>>
>>> *
>>>
>>> **** JEOL Probe Users Listserver
>>>
>>> Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe
>>> Lab, University of Minnesota
>>>
>>> Post a message: send your message to [log in to unmask]
>>>
>>> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
>>>
>>> On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html
>>>
>>> *
>>>
>>>
>>> ********************************************************************
>>> This message does not originate from a known Department of Energy email
>>> system.
>>> Use caution if this message contains attachments, links or requests for
>>> information.
>>>
>>> ********************************************************************
>>>
>>> **** JEOL Probe Users Listserver
>>>
>>> Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe
>>> Lab, University of Minnesota
>>>
>>> Post a message: send your message to [log in to unmask]
>>>
>>> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
>>>
>>> On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html
>>>
>>> *
>>>
>>> **** JEOL Probe Users Listserver
>>>
>>> Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe
>>> Lab, University of Minnesota
>>>
>>> Post a message: send your message to [log in to unmask]
>>>
>>> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
>>>
>>> On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html
>>>
>>> *
>>>
>>>
>>> ********************************************************************
>>> This message does not originate from a known Department of Energy email
>>> system.
>>> Use caution if this message contains attachments, links or requests for
>>> information.
>>>
>>> ********************************************************************
>>> **** JEOL Probe Users Listserver
>>>
>>> Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe
>>> Lab, University of Minnesota
>>>
>>> Post a message: send your message to [log in to unmask]
>>>
>>> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
>>>
>>> On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html
>>>
>>> *
>>>
>> **** JEOL Probe Users Listserver
>>
>> Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe
>> Lab, University of Minnesota
>>
>> Post a message: send your message to [log in to unmask]
>>
>> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
>>
>> On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html
>>
>> *
>>
>
****