Steve,
Sometimes knowing what not to buy helps. So along that line of reasoning...
When I set up our JEOL 8600 I installed a Sepelco Rotameter. The unit I purchased was Cat # 23325. The description is:
“flow range: 0-110 mL/min, with needle valve”
I have found that at the flow rate I use for my two GFPC spectrometers, the rotameter barely registers!
Perhaps Suplco’s 23324, which is a 0-33 mL/Min unit with needle valve, would have been the better choice???
BTW, I use Pfeiffer TL 011 turbo pump oil in my bubble flow indicator at the exit point of the P-10 gas in my system. The slight yellow tint of the oil makes it very easy to see the bubbles, it does not evaporate away very quickly, and I see no trace of it backstreaming towards my detectors.
Mike
On 6/1/10 6:06 AM, "Steven Cogswell" <[log in to unmask]> wrote:
JEOL Probe Users Listserver
Moderator: Ellery Frahm, [log in to unmask], Electron Microprobe Lab, University of Minnesota
Post a message: send your message to [log in to unmask]
Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html
* Hello;
I'd like to clarify my position in the question I asked yesterday.
I have no intention of getting rid of the bubble device, and I have no intention of getting rid of P10. I use the system now with a regulator, a needle valve, and the bubble meter. My needle valve is unmarked as to range, and is really too coarse for fine control over the rate - with the side effect that if you bump the needle valve knob the flow rate changes a lot. My intention is to replace this with a finer control rotameter which has the needle control valve and a visual sight gauge to let you know that it's actually set correct. Since the JEOL spec doesn't actually say what the flow rate is in l/min (or at least I can't find it) I figured I would ask if anyone knew.
Have a nice day.
Best regards,
Steven Cogswell
********************************************************************
Michael M. Cheatham
312 Heroy Geology Laboratory Phone (315)-443-1261
Syracuse University Fax (315)-443-3363
Syracuse, NY 13244-1070
email:[log in to unmask]
http://earthsciences.syr.edu
owner of PLASMACHEM-L: http://listserv.syr.edu/archives/plasmachem-l.html
owner of XRF-L: http://listserv.syr.edu/archives/xrf-l.html
owner of TIMS-L: http://listserv.syr.edu/archives/tims-l.html
********************************************************************
|