PROBEUSERS Archives

June 2009

PROBEUSERS@LISTS.UMN.EDU

Options: Use Monospaced Font
Show HTML Part by Default
Show All Mail Headers

Message: [<< First] [< Prev] [Next >] [Last >>]
Topic: [<< First] [< Prev] [Next >] [Last >>]
Author: [<< First] [< Prev] [Next >] [Last >>]

Print Reply
Subject:
From:
"Wisner, Clarissa A." <[log in to unmask]>
Reply To:
JEOL-Focused Probe Users List <[log in to unmask]>
Date:
Wed, 24 Jun 2009 16:26:54 -0500
Content-Type:
multipart/alternative
Parts/Attachments:
text/plain (6 kB) , text/html (17 kB)
JEOL Probe Users Listserver

Moderator: Ellery Frahm, [log in to unmask],
Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html



*

Hi,

 

I would like to know if anybody with a microprobe, especially a JEOL
733, will be attending the M&M conference in Richmond and would they
like to get together for a discussion?

 

thanks

 

 

Clarissa Wisner
SEM Specialist
G6 MRC
Missouri University of Science and Technology
Rolla, Mo  65409
[log in to unmask]
573-341-4393

'Life isn't about how to survive the storm,
But how to dance in the rain.'

________________________________

From: JEOL-Focused Probe Users List [mailto:[log in to unmask]] On
Behalf Of Ron Rasch
Sent: Friday, May 22, 2009 2:46 AM
To: [log in to unmask]
Subject: Re: [PROBE-USERS] JEOL 733 continuing viability

 

Hi Everybody,

 

Thankyou for your kind help. To quickly address a few issues.

 

Spectrometer Odometer - we don't have them. Sorry this is an 8800 EPMA,
I was answering Ritchie's question about the 733 (which some how did not
make it through) when I thought out loud about my old 8800.

 

Beam Position and charging in OL pole piece - good point, I have seen
this before in other machines, but I do not see any sign of it now.
Image is stable even under high beam currents, and the optical image of
my spot on a fluorescence specimen does not jump about or shift.

 

WDS belt linearity/tension/length - we have a very experienced factory
engineer and this is always checked with belt replacement, plus the new
probe with the same engineer is doing great. Also, I have used
unadjusted belts before and they do not give me this degree of trouble.

 

Backlash issues - Both probes old and new use the same software and same
backlash correction, only the old probe gives trouble, plus the old
probe worked fine with this backlash correction when new. Also, when I
did my reproducibility tests I tried starting numerous times from above
the peak and below the peak, no real difference. Having said that, maybe
it would be a good idea to increase this backlash correction on the
older machine, thanks!

 

I've also tried these various tests with open slits and closed slits,
again no real improvement. 

 

Thanks again for all your help.

 

Best Regards,

Ron

 

________________________________

From: JEOL-Focused Probe Users List on behalf of William J Mushock
Sent: Thu 21/05/2009 23:24
To: [log in to unmask]
Subject: Re: [PROBE-USERS] JEOL 733 continuing viability

JEOL Probe Users Listserver 

Moderator: Ellery Frahm, [log in to unmask], Electron Microprobe Lab,
University of Minnesota 

Post a message: send your message to [log in to unmask] 

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask] 

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html 

* Ron,

First, your software should be setup so the spectrometer approaches the
peak from the same direction
to avoid backlash issues. 

There is a spectrometer adjustment that has a large impact on
reproducibility called the belt linearity adjust. This usually is not a
problem unless the stainless steel belt on the spectrometer has been
replaced. I know some service guys will avoid doing it because it is a
little tricky and very time consuming. When the adjustment is out it is
most difficult to reproduce at the higher L values (>170mm).

You can do a quick check of the belt linearity by defocusing your sample
and re-peaking the spectrometer (at a high spectrometer L-value). First,
defocus the optical image (Z) at +/-20um intervals and re-peak the
spectrometer. If the count rate goes up and is considerably higher with
the sample defocused vs focused, you have a belt linearity problem. The
specification is to have maximum count rate within 40um of sample focus.
If you plot this over the  range of the spectrometer it can also tell
you if your baseplate and crystal tilt are aligned correctly as well.
Note: this check and/or adjustment should be done with at least the
500um slit in place.

Bill

On Thu, May 21, 2009 at 8:19 AM, Ellery Frahm <[log in to unmask]> wrote:

JEOL Probe Users Listserver

Moderator: Ellery Frahm, [log in to unmask],
Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html



*

Hi Ron,

What about the beam position on the sample?  Is that stable?  If, for
example, there is charging in the objective lens pole piece or somewhere
else in the column, the beam could wonder a lot across the sample,
constantly changing the Bragg angles (and therefore L values) and
affecting all the spectrometers.

Best,
Ellery 




On May 20, 2009, at 11:49 PM, Ron Rasch wrote:

Hi Ellery,

Yes all good questions. To go into more detail:

This probe has seen a lot of work, 15 years of 24 hrs a day, 7 days
week, excluding down time etc.
Doing a peak search during Qnt analysis, does not help anymore, but it
use to when it started "going off". For most of the probe's life I only
did a peak search during Standard analysis.
All four WDS spectrometers are giving bad results but to a greater or
lesser extent.
My FCS is probably the best and my H type is probably the worst.
The beam current looks very stable.
The standards are good and have been tested in my new probe. Plus they
were good at the start (15 years ago).
There seems to be no strong connection between L value and
reproducibility, some days the high values some days the low, but mostly
the low L_values are better.
I've had the factory check the base plate, align the crystal tilt and do
the SCA conditions, twice.
I do not believe it's the x-ray detectors (see below), but just in case
I've replaced one Xe detector and the window for one GPC, no great
improvement.

My usual test (apart from just doing a typical analysis) is to do a peak
search on a standard, then take a 10 second count three times, then move
off the peak then repeat.

On any given peak search, the three count rates are quite consistent
with each other, just not consistent with the count rates after the next
peak search. Two very close peak positions may not yield similar count
rates.

Next on my list is to try swapping a stepper motor and gear box with my
new probe, but I'm also looking for any other ideas or things to check.

Cheers,
Ron




-- 