Mime-Version: |
1.0 |
Sender: |
|
Subject: |
|
From: |
|
Date: |
Wed, 11 Jun 2008 10:08:14 -0700 |
Content-Type: |
text/plain; charset="us-ascii"; format=flowed |
Reply-To: |
|
Parts/Attachments: |
|
|
JEOL Probe Users Listserver
Moderator: Ellery Frahm, [log in to unmask],
Electron Microprobe Lab, University of Minnesota
Post a message: send your message to [log in to unmask]
Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]
On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html
*
Hi Ellery,
Two things- I note that the image of the analytical results I posted
previously did not come through the last email to the list server, so
I'm sending it as an attachment in case that works better.
Second, something very interesting. I originally calculated the H2O
results for the Withers glasses two ways. One just using water by
difference from 100% (not measuring oxygen) and also by measuring
oxygen to calculate the H2O from excess measured oxygen. Just as you
did. When I posted the earlier analysis (now attached), I noticed
that it was still in the mode where H2O was being calculated by
stoichiometry to excess oxygen and included in the matrix correction.
So then when I wanted to show you the effect on SiO2 from not
including H2O, I simply unchecked the stoichiometry to excess oxygen
option and reported the change in SiO2 concentration due to the
change in the matrix correction. That wasn't quite right. Because I
was still measuring oxygen (and including it in the matrix
correction), the concentration change in SiO2 I reported to you was
only due to the lack of hydrogen in the matrix!
I would never have thought that 0.59% hydrogen could cause such a
large matrix effect in the SiO2 concentration. I'm going to double
check this but here is a preliminary summary:
SiO2
without H2O 70.025
with excess O 70 22 (but without the 0.59% H)
with H2O 70.57
Note that in the last step I calculated H2O two different ways just
to double check. First using the hydrogen by stoichiometry to excess
measured oxygen calculation and also by calculating oxygen by
stoichiometry and adding 0.59% hydrogen to the matrix as a fixed
concentration. The SiO2 results are almost exactly the same for both methods.
I don't quite believe it but it appears that the matrix effect on
SiO2 of 0.59% hydrogen actually a larger effect than the 4.9% excess
oxygen (for this composition anyway).
I just checked again- it seems to be. The Z correction on Si ka for
no hydrogen goes from 1.2194 to 1.2254 with 0.59% hydrogen. That's a
0.48% change in the Si ka intensity. Weird.
john
John J. Donovan
Director, MicroAnalytical Facility
CAMCOR, University of Oregon
(541) 346-4632 (office)
(541) 346-4655 (lab)
www.epmalab.uoregon.edu
|
|
|