Hi Kyle,
if your transmitted image is free of astigmatism, then the dichroic must be 
responsible for the astigmatism in the reflected path (assuming there are no
other components between the dichoic and the cameras). Try a flat mirror (I 
mean really flat, 6mm substrate, lambda/10 specs) instead of the dichroic 
and see if the astigmatism is gone.

The new 3mm dichroics have similar specs (lambda/4 to lambda/10), so if it 
is mounted properly (it's not about exact angles, but you must not introduce
any stress that may lead to bending), it should work OK. Try another 
dichroic, maybe one factory-mounted in a cube.

Also, the closer the dichroic to the camera, the smaller the effects of 
curvature. I don't think infinty space is less 'sensitive' to astigmatism. 
Quite the opposite - least sensitive should be a position conjugate to image
plane (but then, every piece of dust will be in focus on the camera...).

Best, zdenek





Zdenek Svindrych, Ph.D.

W.M. Keck Center for Cellular Imaging (PLSB 003)

Department of Biology,University of Virginia

409 McCormick Rd, Charlottesville, VA-22904

http://www.kcci.virginia.edu/

tel: 434-982-4869

Annual FRET Workshop (March 6-10, 2017): http://kcci.virginia.edu/workshop-
2017


---------- Původní zpráva ----------
Od: Kyle Douglass <[log in to unmask]>
Komu: [log in to unmask]
Datum: 13. 2. 2017 9:29:31
Předmět: Reflected image -> astigmatism

"***** 
To join, leave or search the confocal microscopy listserv, go to: 
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy 
Post images on http://www.imgur.com and include the link in your posting. 
***** 

Hi everyone, 

A couple years ago there was a post on this forum about significant 
astigmatism appearing in the reflected path of a custom-built, two 
camera TIRF setup. 
(http://confocal-microscopy-list.588098.n2.nabble.com/Reflected-image-gt-
astigmatism-td7584402.html) 
The original poster had placed a dichroic before a pair of matched tube 
lenses to split the two color channels onto two separate cameras. The 
general consensus was that you could "buy your way out of the problem" 
by buying a thick, ultraflat dichroic. 

I am trying a similar approach on a setup in our lab but, due to space 
constraints that are not easily overcome, have tried first placing the 
channel-splitting dichroic in the image space after the final tube lens 
instead of in the infinity space between the objective and tube lens. 
Even with a 3 mm thick ultraflat dichroic, I see moderately bad 
astigmatism in the reflected channel. I have ruled out other possible 
sources of the astigmatism, like an additional dichroic for the 
excitation light and filters that already lie in the infinity space 
between the objective and tube lens. 

My question is: is it even worth trying to eliminate the astigmatism in 
the reflected path of a split-channel setup if the dichroic does not lie 
in the infinity space? Or is it pretty much always a bad idea to put the 
splitter dichroic in the image space? 

My current suspicion is that the tolerances in the alignment have to be 
very, very tight to avoid astigmatism with the splitter in the image space. 

Thanks for your feedback. 

Kyle 

-- 
Kyle M. Douglass, PhD 
Post-doctoral researcher 
The Laboratory of Experimental Biophysics 
EPFL, Lausanne, Switzerland 
http://kmdouglass.github.io 