JEOL Probe Users Listserver Moderator: Ellery Frahm, [log in to unmask], Electron Microprobe Lab, University of Minnesota Post a message: send your message to [log in to unmask] Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask] On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html * I would retain the bubble device. You have to remember that the flow rate is not critical, all you are doing is making sure that you are putting sufficient P10 into the flow counters to compensate for leakage/diffusion through the very thin (and somewhat porous) polymer window. The only role of the outflow is to give you assurance that you are putting in at least slightly more than is diffusing out, so the bubble meters are ideal for the purpose. You want the bare minimum flow to achieve this, as P10 is really expensive. I doubt whether you will find a rotameter capable of registering this bare minimum flow. The bubble "meters" also have the advantage of preventing back diffusion of air into the flow counters should the flow stop eg if the cylinder runs out, or if someone inadvertantly turns the flow off. Incidentally, for a stable flow the regulator output pressure should be appreciable (say 10 to 20psi), and the flow adjustment done with a needle valve (a restrictor) downstream of the pressure regulator. Stable flow rates will not be achieved by omitting the needle valve restrictor and attempting to adjust the rate with the pressure regulator alone. cheers Ritchie Sims On 1 Jun 2010 at 3:46, M.Piranian wrote: JEOL Probe Users Listserver Moderator: Ellery Frahm, [log in to unmask], Electron Microprobe Lab, University of Minnesota Post a message: send your message to [log in to unmask] Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask] On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html * Wouldn't it be easier and more accurate to take the tank volume, divide by the number of P10 spectrometers and the time the tank lasts. That eliminates bubble size uncertainty. mp --- On Mon, 5/31/10, Straszheim, Warren E [M S E] <[log in to unmask]> wrote: From: Straszheim, Warren E [M S E] <[log in to unmask]> Subject: Re: [PROBE-USERS] What's your P10 flow rate range? To: [log in to unmask] Date: Monday, May 31, 2010, 9:53 AM JEOL Probe Users Listserver Moderator: Ellery Frahm, [log in to unmask], Electron Microprobe Lab, University of Minnesota Post a message: send your message to [log in to unmask] Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask] On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html * FWIW, I recall 20 drops per milliliter for water. I don't how bubbles correspond to ml. I would set the flow and run it into an inverted, water-filled graduated cylinder. Supposing a bubble is equivalent to a drop, then 2 bubbles per second would be 120 per minute or 6 ml/min. That ain't much. I would doubt that you could find a rotameter reading that low, but I could be wrong. Personally, the tube in liquid seems good and simple and cheap enough for me. Warren S. From: JEOL-Focused Probe Users List [[log in to unmask]] On Behalf Of Steven Cogswell [[log in to unmask]] Sent: Monday, May 31, 2010 8:27 AM To: [log in to unmask] Subject: [PROBE-USERS] What's your P10 flow rate range? JEOL Probe Users Listserver Moderator: Ellery Frahm, [log in to unmask], Electron Microprobe Lab, University of Minnesota Post a message: send your message to [log in to unmask] Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask] On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html * Hi Probe folks; I'm back again, but amazingly this time it's not because I have a technical problem. This time it's just an easy question. Our JEOL 733 uses some P10 gas-flow spectrometers. We have the P10 tank hooked up to it and use one of those little "flow meters", as JEOL calls them, to see what the flow rate through the spectrometer is. Manual says "1-2 bubbles per second" when the little glass thingie (technical term) has alcohol in it. I'm ambitious to replace the valve we use now to control the flow (very coarse, very crude) with an actual useful rotameter. Problem is, I have no clue what range I should buy. If I creak back to my old chemistry days usually they told you things like "1 drop is about 1 mL". Rotameters are expensive and I'd like to buy one with a correct range and not be disappointed. So for folks who actually know what the gas flow rate is, what are you using? Best regards, Steven Cogswell