Hi Peter,

I agree with what Ron said.  We have an 8200 which recently tripped out with an oil warning when it was actually the water chiller that had failed.  

The oil level in the pump only needs to drop slightly for it to trip out but it is simple to top up and should be about an inch from the top of the reservoir.  We were given some type of Shell diala oil by the Jeol engineer.

Cheers,

Joe
--------------------------------------------------------------------------------------

Joe Pickles
Senior Technician, Camborne School of Mines

Tel:+44 (0)1326 371851

Daphne du Maurier Building 3.125, University of Exeter, Penryn Campus, Penryn, TR10 9FE, Cornwall.

On 26 Sep 2014, at 03:00, Ron Rasch <[log in to unmask]> wrote:

Hi Peter, 

We have a JEOL 8200 EPMA. Our lens oil is cooled by the same cold water that cools the DP's. If the cooling water gets too warm, and the oil gets into the 30's (deg C) the the probe will switch itself off.
So check you cooling water temp first.

It's rare for the oil level to drop without a obvious leak and oily patch on the floor. Ours is almost 10 years old and the level is still fine. Our old 8800 blew an oil hose once, that was messy.

If you open the box that contains the oil pump and reservoir, three screws (from memory) allow you to take the top off and check the oil level in the reservoir (best done with the probe off). 

If you stand clear of any moving parts on the pump, and keep clear of any electrical connections, you can hold the lid up a little, keep the return line in the lid pointed at the reservoir, and move the float switch up and down with your fingers to test if this is the problem. It's quite easy, but messy if you don't keep the return flow pointed into the reservoir. Make sure you feel confident to safely do it by first examining the system when off. I always have a second person in the room at the power switch to turn it all off if I slip or drop anything. I wear nitrile gloves and have spare oil (and paper towels) on hand. Safety first !

A level drop of about an inch will cause problems. We top up with transformer oil, I use Shell Diala B. Simply because when I needed 1 L, I went to the high voltage lab of the electrical engineering department, and that is what they use. They buy it by the 200L drum, but I only need 1L for top ups, which they happily gave me.

I hope this helps.

Cheers,
Ron


On 26/09/2014, at 6:24 AM, "Straszheim, Warren E [BIOTC]" <[log in to unmask]>
 wrote:

I will be seeing our engineer tomorrow. I may be able to relay your questions on to him. I do not know the answers offhand.
 
I think we had a similar problem a few years back when our 8200 was about six years old. I think I was concerned about a whine from the pump. In investigating that, he found that the oil level was just a little bit low for the safety circuit to be happy. There was really plenty of oil present, and I think the engineer bypassed the safety TEMPORARILY to allow debugging. He topped off the tank before he was finished. I recall it being a challenge trying to get the oil into the system. I seem to recall only a small hole.
 
We have a gallon of transformer oil around here. I suspect it is what they used. JEOL has recently fitted out our system with a flowmeter and thermocouple to prevent failure of the objective lens due to a loss of oil. Ironically, this came about a year after we lost our objective lens maybe due to a loss of cooling.
 
Warren
 
From: JEOL-Focused Probe Users List [mailto:PROBEUSERS@LISTS.UMN.EDU] On Behalf Of Michael, Peter
Sent: Thursday, September 25, 2014 3:08 PM
To: [log in to unmask]
Subject: [PROBEUSERS] oil for lens in JEOL 8200
 
Dear JEOL users:
 
Our JEOL 8200 is tripping its power off and giving us an error light on panel for the oil that is used to cool the lens and some electronics.  I have some questions: 
 
Does anyone know how high the oil must be in the oil reservoir for the pump sensor not to trip off the pump?  (There is a float switch). 
 
Is it common / uncommon / or rare / for the oil elvel to drop slightly over time?  
 
Does anyone know what type of oil it is?
 
And, where it can be obtained?  
 
Thanks,
 
 
Peter
 
 
 
Peter Michael
Department of Geosciences
The University of Tulsa
800 S. Tucker Avenue
Tulsa, OK  74104
918-631-3017
fax: 918-631-2091
[log in to unmask]


 
**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html

*

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html



*


**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html



*


**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/listserver.html

*