Tried but still the dark part at the left of the image remains.

Στις Πέμ, 3 Ιαν 2019, 21:34 ο χρήστης Cyril Kirkland <[log in to unmask]> έγραψε:
If it is clipping at low mag try adjusting the Objective aperture. That is usually the first culprit.



On Jan 3, 2019, at 11:21 AM, Straszheim, Warren E [BIOTC] <[log in to unmask]> wrote:

For the recent replies, it seems someone else must have suggested a problem with alignment. The 40x image definitely looks like an alignment issue. Something seems to be clipping the left side of the scan. I would be curious what it looks like as you zoom in gradually. I presume the shading on the left side moves out of the field of view.

 

Others have commented on how to do alignments. I would agree, start with the manual.

 

Do you have a cold finger on that probe? I vaguely recall that they can be rotated in and out of position. Perhaps it is in an intermediate position.

 

You definitely seem to have a problem with linearity in the X-scan on the right side of the image. Something changed markedly between 950x and 1000x. Many systems change circuitry as you increase magnification. I can clearly hear relays kicking in and out on my current SEM. I suppose the probe would be the same. I would think the magnification has to be calibrated for each system and maybe matched between left and right sides and between top and bottom. Maybe you are simply out of calibration. (I don’t know that is something intended for regular users, even super-users.) Maybe it is a problem due to failing electronics. If so, I think you are looking at a service call.

 

We have an 8200 that is no longer under contract. Usage is down and management could not justify $33K/year, nor could I. I had to call in a service visit on-demand. I think it cost me $8000 but that is a lot less than paying for the contract. It does help a lot that we have a service engineer about 50 miles away. It saves on travel costs.

 

Warren Straszheim

Iowa State University

 

From: JEOL-Focused Probe Users List [mailto:[log in to unmask]] On Behalf Of Ioannis Baziotis
Sent: Thursday, January 03, 2019 10:47 AM
To: [log in to unmask]
Subject: Re: [PROBEUSERS] [EXTERNAL] Re: [PROBEUSERS] Re: [PROBEUSERS] Re: [PROBEUSERS] URGENT HELP: JEOL JXA 8900 Superprobe

 

OK, here are the photos.

----------------------------------------------------------------
Ioannis Baziotis [log in to unmask]
Assistant Professor of Mineralogy and Petrology
Agricultural University of Athens
Laboratory of Mineralogy and Geology
Athens, 11855 Greece
(+30)210-529-4155
Twitter: https://twitter.com/EGU_GMPV
Personal site: http://ibaziotis.wix.com/ioannis-baziotis

----------------------------------------------------------------

Next dream is "on-the-go" .........mid -April 2019

 

 

 

Στις Πέμ, 3 Ιαν 2019 στις 6:42 μ.μ., ο/η Ioannis Baziotis <[log in to unmask]> έγραψε:

Here is a set of consecutive photos (starting from low mag X40).  As you see, part of the image (at left) is dark (the same occurs both in SE and BSE). Also, toward higher mag, the image becomes "blurry" (or something like that). 

----------------------------------------------------------------
Ioannis Baziotis [log in to unmask]
Assistant Professor of Mineralogy and Petrology
Agricultural University of Athens
Laboratory of Mineralogy and Geology
Athens, 11855 Greece
(+30)210-529-4155
Twitter: https://twitter.com/EGU_GMPV
Personal site: http://ibaziotis.wix.com/ioannis-baziotis

----------------------------------------------------------------

Next dream is "on-the-go" .........mid -April 2019

 

 

 

Στις Πέμ, 3 Ιαν 2019 στις 6:35 μ.μ., ο/η William J Mushock <[log in to unmask]> έγραψε:

Could you provide a picture or video of what you describe as internal reflections? And, are you convinced this is the result

of the same issue? Confession: In a former life I was a field service rep for JEOL and had a lot of experience on the 8900s.

I think it would be a big help to see what you are describing.

 

Bill

 

On Thu, Jan 3, 2019 at 10:01 AM Collins, Keith W. <[log in to unmask]> wrote:

Unfortunately, It may be time to schedule a service visit from JEOL.

 

From: JEOL-Focused Probe Users List <[log in to unmask]> On Behalf Of Ioannis Baziotis
Sent: Thursday, January 3, 2019 3:55 AM
To: [log in to unmask]
Subject: [EXTERNAL] Re: [PROBEUSERS] Re: [PROBEUSERS] Re: [PROBEUSERS] URGENT HELP: JEOL JXA 8900 Superprobe

 

We have cleaned and reseat all pc boards. Checked all the fuses at the front and back of the control unit but nothing changed. Any other ideas to test?

 

Στις Παρ, 28 Δεκ 2018, 21:30 ο χρήστης Collins, Keith W. <[log in to unmask]> έγραψε:

You can generally visually look at the fuse and see if the wire has burned out (magnifying glass helps) or use a volt ohm meter to test,.  IF the fuse is bad you replace it with the same rated fuse.  Fuses may be located in a block of fuses, on the circuit board or in a fuse holder in a wire (usually a twist to open).   Fuses should be marked on the wiring diagram.   I would suggest a service call.

 

From: JEOL-Focused Probe Users List <[log in to unmask]> On Behalf Of Ioannis Baziotis
Sent: Friday, December 28, 2018 10:56 AM
To: [log in to unmask]
Subject: [EXTERNAL] Re: [PROBEUSERS] Re: [PROBEUSERS] URGENT HELP: JEOL JXA 8900 Superprobe

 

Ok. 

1) how do you check them?

2) how do you replace them?

 

Στις Παρ, 28 Δεκ 2018, 20:37 ο χρήστης Collins, Keith W. <[log in to unmask]> έγραψε:

Also check fuses a blown fuse to the scan coils can result in half an image.

 

From: JEOL-Focused Probe Users List <[log in to unmask]> On Behalf Of Ioannis Baziotis
Sent: Friday, December 28, 2018 9:22 AM
To: [log in to unmask]
Subject: [EXTERNAL] Re: [PROBEUSERS] URGENT HELP: JEOL JXA 8900 Superprobe

 

Great...i`ll try it. so, which PC boards you exactly mean? The one seeting at the front part of the control unit, or at the back, too?

----------------------------------------------------------------
Ioannis Baziotis [log in to unmask]
Assistant Professor of Mineralogy and Petrology
Agricultural University of Athens
Laboratory of Mineralogy and Geology
Athens, 11855 Greece
(+30)210-529-4155
Twitter: https://twitter.com/EGU_GMPV
Personal site: http://ibaziotis.wix.com/ioannis-baziotis

----------------------------------------------------------------

Next dream is "on-the-go" .........mid -April 2019

 

 

 

https://ipmcdn.avast.com/images/icons/icon-envelope-tick-round-orange-animated-no-repeat-v1.gif

Απαλλαγμένο από ιούς. www.avast.com

 

Στις Παρ, 28 Δεκ 2018 στις 5:56 μ.μ., ο/η William J Mushock <[log in to unmask]> έγραψε:

Ioannis,

 

Turn off console power, remove the lower front cover and try re-seating the PC boards. Sounds like you may have developed

a bad contact (oxidation) on the edge connector of one of the boards.

 

 

On Fri, Dec 28, 2018 at 8:25 AM Ioannis Baziotis <[log in to unmask]> wrote:

Dear Colleagues,

I have noticed the following problem on my JEOL JXA-8900 Superprobe.
At the lowest magnification (40X), the 30% of the screen (left part) is black (either under SE or BSE). Please note, that under OM (reflected light) the image looks perfect. Also, at higher mag the horizontal movement (when inspecting a sample) is like having internal reflections. The filament has been replaced very recently, however, i had the same problem still occurring since the last week with the previous filament. Also, at higher mag (from 950X to 1000X and 9500X to 10000X) there is a considerable shift on horizontal axis. I tried the objective lense aperture (X and Y) however, nothing changed. Any idea for the problem and how can i fix it?

All the best.

****
JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask],
Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*


********************************************************************
This message does not originate from a known Department of Energy email system.
Use caution if this message contains attachments, links or requests for information.

********************************************************************

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*


********************************************************************
This message does not originate from a known Department of Energy email system.
Use caution if this message contains attachments, links or requests for information.

********************************************************************

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*


********************************************************************
This message does not originate from a known Department of Energy email system.
Use caution if this message contains attachments, links or requests for information.

********************************************************************

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*

**** JEOL Probe Users Listserver

Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota

Post a message: send your message to [log in to unmask]

Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask]

On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html

*