Julien, FYI, the charging could also be the result of a poor connection somewhere between your sample and the ammeter. Doesn't have to be the sample. A cold solder joint or a dirty contact etc... Best of Luck, Bill On Tue, Jul 14, 2020 at 1:50 PM Allaz Julien <[log in to unmask]> wrote: > Thanks for the additional suggestion! > > The W filament is actually pretty new with less than 100 hours, and I’ve > seen this behaviour of change in count rate with beam current (not purely > related to a bad DT correction) with other filament. > > We have the LaB6 option, and I have yet to load such a crystal on this > EPMA. We just received a bunch and I will likely load it after the next > maintenance. > > We have the JEOL PC-SEM and PC-EPMA on Windows 10 with the following > versions: > - PC-SEM 3.0.1.26 > - PC-EPMA 1.16.0.2 > - EOS sub system 1.15.3.0 > > I believe these where the version accessible to JEOL Germany as of April > 2019 (update at installation). > > So far it’s running smoothly beside some known software and other minor > hardware issues. However, I should say that I prefer Probe for EPMA for > most of my analysis and mapping, and only use the JEOL software for beam > alignment, imaging, combined EDS-WDS mapping, and leave most of my quant > analysis and complex cases to PfE. However, all test I sent so far were > exclusively run with the JEOL software. > > I’ll try to hook my ammeter we have at the SEM later this week if I have > time (and when the SEM is not in use)… > > I’ll also check the DH limit on the EDS at high current, although I have > observed this issue in many samples (test shown here made on our permanent > standard block, other sample tested are several thin sections freshly > coated). > > Best, > > J. > > > On 14 Jul 2020, at 19:31, William J Mushock <[log in to unmask]> > wrote: > > Hi Julien, > > Is it possible your sample is building a charge at the higher currents. If > you have an EDS detector you > might be able to verify this by monitoring the Dwayne-Hunt limit. > > Cheers, > Bill > > On Thu, Jul 9, 2020 at 11:44 AM Allaz Julien <[log in to unmask]> > wrote: > >> Dear all, >> >> I hope you are all safe and healthy… >> >> Ever since JEOL installed our new 8230 last year, I’m having issues with >> the count rate at low vs. high beam current. At first, I thought this was a >> dead time issue (still set to the original value of 1.1 µs), but I can >> ensure you it is not... >> >> My most recent test was done from 10 nA to 200 nA, looking at a LOW >> QUANTITY element (Ti in hornblende ~1.4 wt% TiO2). With this test, even the >> highest beam current yields a relatively low count rate (around 6000 to >> 8000 counts per second), and therefore I can almost for sure eliminate a >> dead time issue. As you can see from the attached PDF, ALL five >> spectrometers show the same behaviour, with an increase in count rate with >> beam current, around 4 to 6% at 70 nA (reference = measure done at 10 nA), >> and a whooping 8 to 12% at 100 and 200 nA!!! >> >> This result let me think that there is a non-linearity in the beam >> current measurement (i.e., when the instrument measures 200 nA with the >> Faraday cup, it is in reality 220 or 225 nA…). Or maybe something else??? >> >> Did anyone observe such a behaviour on their 8230 or 8530? Is it possible >> to adjust this, and if so, how? >> >> I know that Cameca acknowledge this issue; in their case, the “jump” >> occurs around 50 nA (and there is possibly another “jump” at much lower >> current; these jumps are due to different “loops” measuring either a low, >> medium, or high current or something like this...). Cameca “solves” this >> issue by adding to the dead time correction a linearity coefficient that is >> applied beyond a certain current threshold (this factor can be found in the >> dead time setting of Peak Sight). Does JEOL has the same thing somewhere >> (maybe not accessible unless logged in as a JEOL engineer)??? >> >> Best, >> >> Julien >> >> >> >> ########################### >> >> *Dr. Julien Allaz *Head assistant for SEM/EPMA >> Inst. für Geochemie und Petrologie >> ETH Zürich >> NW E 84 >> Clausiusstrasse 25 >> 8092 Zürich >> Switzerland >> >> Tel: +41 44 632 37 20 >> Fax: +41 44 632 16 36 >> Email: [log in to unmask] <[log in to unmask]> >> ########################### >> >> **** JEOL Probe Users Listserver >> >> Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe >> Lab, University of Minnesota >> >> Post a message: send your message to [log in to unmask] >> >> Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask] >> >> On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html >> >> * >> > **** JEOL Probe Users Listserver > > Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe > Lab, University of Minnesota > > Post a message: send your message to [log in to unmask] > > Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask] > > On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html > > * > > > **** JEOL Probe Users Listserver > > Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe > Lab, University of Minnesota > > Post a message: send your message to [log in to unmask] > > Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask] > > On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html > > * > **** JEOL Probe Users Listserver Moderator: Anette von der Handt, [log in to unmask], Electron Microprobe Lab, University of Minnesota Post a message: send your message to [log in to unmask] Unsubscribe: send "SIGNOFF PROBEUSERS" to [log in to unmask] On-line help and FAQ: http://probelab.geo.umn.edu/jeoluserlist.html *