Mark,
By my reckoning it's after 4 in the morning in Auckland - what are
you doing replying at this hour? (By the way, I'm in Taiwan, not Oz).
Guy
Optical Imaging Techniques in Cell Biology
by Guy Cox CRC Press / Taylor & Francis
http://www.guycox.com/optical.htm
______________________________________________
Associate Professor Guy Cox, MA, DPhil(Oxon)
Australian Centre for Microscopy & Microanalysis,
Madsen Building F09, University of Sydney, NSW 2006
Phone +61 2 9351 3176 Fax +61 2 9351 7682
Mobile 0413 281 861
______________________________________________
http://www.guycox.net
-----Original Message-----
From: Confocal Microscopy List [mailto:[log in to unmask]]
On Behalf Of Mark Cannell
Sent: Friday, 17 June 2011 2:41 AM
To: [log in to unmask]
Subject: Re: averaging vs. accumulation for noise reduction - is there a
difference?
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****
It also depends on how the readout bandwidth is controlled for
different scan speeds...
Cheers
On 16/06/2011, at 5:26 PM, Stanislav Vitha wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hallo,
> this is a very basic question, but I cannot figure this out from what
I have
> been reading, so a simple explanation for a non-physicist would be
much
> appreciated:
>
> Is there a real difference in the improvement of the signal to noise
ratio
> between frame averaging (or accumulation) and longer dwell times
(slower
> scan) for a point-scanning confocal witrh a PMT detector?
>
> For instance, using single point scanning confocal, 12-bit
acquisition.
>
> a) averaging (or accumulating) 5 frames, 4 microseconds per pixel
> b) acquiring a single frame, 20 microseconds per pixel
>
> Assumptions:
> no saturation of the detector;
> stable environmental conditions, no focus drift, etc
>
> Would it matter (for the dfference between the two scenarios) if it
was analog
> detection or photon counting detection?
>
> I will run this little test later, but I am curious what you think.
>
> I thought that at least for the photon counting mode, the two
important
> factors would be the dark counts and the total number of counts
detected, so
> whether it is acquired in one scan or in 5 scans, it should be the
same. My
> camera expert here insists that the averaging scheme will give better
noise
> suppression.
>
> Thanks!
>
>
> Stan Vitha
>
> Microscopy and Imaging Center
> Texas A&M University
-----