CONFOCALMICROSCOPY Archives

October 2014

CONFOCALMICROSCOPY@LISTS.UMN.EDU

Options: Use Monospaced Font
Show Text Part by Default
Condense Mail Headers

Message: [<< First] [< Prev] [Next >] [Last >>]
Topic: [<< First] [< Prev] [Next >] [Last >>]
Author: [<< First] [< Prev] [Next >] [Last >>]

Print Reply
Content-Transfer-Encoding:
quoted-printable
Sender:
Confocal Microscopy List <[log in to unmask]>
Subject:
From:
"Claire Brown, Dr." <[log in to unmask]>
Date:
Mon, 20 Oct 2014 15:00:25 +0000
Content-Type:
text/plain; charset="iso-8859-1"
MIME-Version:
1.0
Reply-To:
Confocal Microscopy List <[log in to unmask]>
Parts/Attachments:
text/plain (140 lines)
*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
Post images on http://www.imgur.com and include the link in your posting.
*****

Hello All,

The ABRF Light Microscopy Research Group (LMRG) has spent considerable time designing, testing and refining the sample type, preparation and imaging protocols for our latest study. The new study will focus on metrics for spherical aberration, sensitivity and signal-to-noise for 3D imaging using a microsphere sample in a tissue like environment. The study overview is below. If you register for the study you will receive an imaging protocol and a sample - which MUST be imaged within one week of its arrival - NO ADDITIONAL SAMPLES WILL BE SENT - so we are asking you to identify an ideal time frame to receive the sample and collect the data.

Due to the data intensive nature of this current study we are only accepting 100 participants. Please fill in the survey by October 29, 2014.

https://www.surveymonkey.com/r/LMRG_Study_3

FYI aside from a pre-meeting image processing and analysis workshop, a keynote microscopy talk (John Condeelis - Albert Einstein) and a full three day microscopy track the ABRF 2015 meeting in St. Louis, MO March 28-31 will feature a session on microscope standards! For more info see conf.abrf.org

If you have any questions or concerns feel free to email me.

Sincerely,

Claire


ABRF-LMRG Study Overview

Introduction: The third study of the Association of Biomolecular Resource Facilities (ABRF) Light Microscopy Research Group (LMRG) is aimed at creating a 3D biologically relevant test slide and imaging protocol to test for 1) system resolution and distortions in 2D and 3D, 2) the dependence of Intensity quantification and image signal-to-noise of the microscope on imaging depth and 3) the dependence of the microscope sensitivity on imaging depth.

Specimen Detail: Fluorescence microspheres (Life Technologies) imbedded in a 120 µm-thick layer of CyGel (BioStatus, UK, Cat# Cy10500) with a refractive index of 1.36 closely match biological tissue. Double-sided adhesive 18 mm square spacers with a well (9 mm diameter, 120 µm deep) were used for the sample preparation (Electron Microscopy Sciences, Cat# 70327-8s).

Microspheres:
Diameter

Relative Intensity

Colour

Excitation Maxima

Emission Maxima

Catalog #

1.0 µm

100%

Red

580

605

F13083

2.5 µm

10%

Green

488

515

L14821

2.5 µm

100%

Green

488

515

L14821

6.0 µm

20%

Deep Red

633

660

L14819

6.0 µm

100%

Deep Red

633

660

L14819

15 µm

100%

Blue Core
Orange Ring

365
560

430
580

F7236


Imaging Conditions: The microsphere samples should be imaged from the coverslip up to 100 µm into the sample. The sample region should include a good selection of 5-10 of each of the different microspheres listed above. A z-stack of images should be collected as per the attached protocols with a minimal x-y and z resolution 0.35 µm. This will result in a data set with approximately 300 images having all the information required to test system resolution, quantification, the signal-to-noise ratio (S/N), and sensitivity.

Metrics: Image stacks will be sent to the LMRG for image analysis, consolidation of the data and interpretation of the results. Briefly, here are the metrics that will be measured.

1)     System Resolution

a.      2D Resolution: PSF FWHM along the x-y axis for 5 microspheres at each depth of 0-25 µm, 25-50 µm, 50-75 µm, 75-100 µm.

b.      3D Resolution: PSF FWHM along the x-z or y-z axis for 5 microspheres at each depth of 0-25 µm, 25-50 µm, 50-75 µm, 75-100 µm.

c.      Visual Inspection of PSF Shape: Visual inspection for aberrations and scoring of PSF shape along the x-z and y-z axes.

2)     Quantification and S/N

a.      Quantify Relative Microsphere Intensities: Measure 5 microsphere pair intensity ratios at each depth of 0-25 µm, 25-50 µm, 50-75 µm, 75-100 µm.

b.      Measurement of S/N ratio: S/N ratio for microsphere intensity of 5 microspheres at each depth of 0-25 µm, 25-50 µm, 50-75 µm, 75-100 µm.

3)     Sensitivity

a.     Microsphere Intensity: Compare the intensity of 5 microspheres at depths of 0-25 *m, 25-50 *m, 50-75 *m, 75-100 *m.