To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****
I could use some of the lists’ collective wisdom/experience on this project.
As part of the next phase of quality assurance testing of light microscopes
(see references at the bottom of email for past tests) the ABRF light
microscopy research group (
<http://www.abrf.org/index.cfm/group.show/LightMicroscopyResearchGroup.54.ht
m>
http://www.abrf.org/index.cfm/group.show/LightMicroscopyResearchGroup.54.htm
) is developing a phantom that can be imaged on different light microscope
platforms (both wide-field and confocal). We are trying to develop
something akin to the phantoms that are used for CAT, PET scans.
The ideal phantom would be; ~100um thick, fluorescent, composed of fibers
of at least three different diameters (~5,25 & 50µM) and have three
different bleach rates ( i.e. couldn’t bleach if you tried, fluorescein and
something in between). Ideally, all samples would have identical 3D
structure, but this is not completely necessary. We are planning on giving
away these samples similar to the last two sets, therefore keeping the cost
down is a necessity.
Here is where I am at:
Micro-fluidics would be great, but unless someone volunteers to take on the
development the $ is too high.
Can use Swiffer fiber (never bleach) and other fluorochromes soaked fibers
(think fibers from filter paper or Kimwipes).
Any suggestions/past experience would be greatly appreciated.
Recent paper from the ABRF light microscopy research group:
Stack, R., Bayles, C., Girard, A., Martin, K., Opansky, C., Schulz, K., and
Cole, R. (2011) Quality Assurance Testing for Modern Optical Imaging
Systems. Microscopy & Microanalysis 17(4):598-606. DOI:
10.1017/S1431927611000237
Cole, R.W., Jinadasa, T., and Brown, C.M. (2011) Resolution and Quality
Control of Confocal Microscopy Optics. Nature Prot. 6 (12): 1929–1941.
doi:10.1038/nprot.2011.407
Richard Cole
Research Scientist V
Director: Advanced Light Microscopy Core Unit
Wadsworth Center
Research Assistant Professor
Dept. of Biomedical Sciences
School of Public Health State University of New York
P.O. Box 509 Albany N.Y. 12201-0509
518-474-7048 Phone
518-474-4430 Fax
Email <mailto:[log in to unmask]> [log in to unmask]
Website <http://www.wadsworth.org/cores/alm/index.htm>
www.wadsworth.org/cores/alm/index.htm
IMPORTANT NOTICE: This e-mail and any attachments may contain
confidential or sensitive information which is, or may be, legally
privileged or otherwise protected by law from further disclosure. It
is intended only for the addressee. If you received this in error or
from someone who was not authorized to send it to you, please do not
distribute, copy or use it or any attachments. Please notify the
sender immediately by reply e-mail and delete this from your
system. Thank you for your cooperation.
|