David Knecht wrote:
> Is the excitation spot size for a laser scanning confocal microscope
> the same as that of an epifluorescence widefield microscope (same
> objective, etc.)? Put another way, if I removed the PMT pinhole,
> would I get an image comparable to a widefield image? Thanks- Dave
If I understand you correctly, the answer is no. For a widefield
microscope, the entire field of view is illuminated at one time. For a
confocal microscope (under ideal circumstances, at least), the area of
excitation is a diffraction-limited spot--i.e. it's sub-micron,
depending on wavelength and how you define the edge.
Martin
--