Dan -
Good point! Thanks for the explanation. I guess, another solution
might be to put the filters at a certain distance and at a slight
angle to each other, if you can afford the slight spectral change.
Beat
At 20:19 12-03-2008, you wrote:
>Beat/ Julio,
> If you are talking strictly about interference filters (made
> with dielectric coatings) the Absorbance of a particular filter is
> mostly only a function of the glass substrate the coating is
> deposited on, as the filter is working to attenuate incoming light
> through reflection at the filter's front surface. If you place two
> of these in series, you will not see an additive blocking effect
> because of reflections bouncing back from the second filter to the
> first, to second, and so on, and there is always some transmittance
> through the second filter that inhibits the level of blocking you
> might otherwise expect to achieve.
>If you were to place a low absorbing piece of glass (such as an NG
>glass with ~95%T) in between the two filters, then the slight
>absorption properties of that glass would inhibit the reflections
>and you would achieve (or nearly so) the additive blocking.
>If you were using absorption glasses only as a blocking component,
>then you would get the full benefit of their OD blocking at a given
>wavelength.
>cheers,