Hi Shanna,
Yes, as Michael mentioned already, you can do a correction with ZEN 2009.
It's called Auto Z Brightness correction and you can find it in the Z stack dialogue window at the bottom. It's linear and is very easy to set up. Instructions are given on page 4-71 of the full manual that I have. I have a copy in PDF so if you don't have this, let me know and I will send it to your email address.