Electronics components are more often all or none in failure modes.
Connectors are a different issue tho'. and more likely to fail. Try re-
plugging in all the connectors you can. Also, I assume you have
checked/tested and proved that it's not the laser illumination side of
things?
Hope this helps
Mark
On 23/10/2010, at 8:19 AM, Sandrine Pouvreau wrote:
> *****
> To join, leave or search the confocal microscopy listserv, go to:
> http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
> *****
>
> Hi all.
> We have a confocal microscope LSM 5 exciter (Zeiss) which is lately
> having
> some troubles. In brief, the fluorescence intensity is changing
> during the
> record, in a line –related pattern (in other word, the intensity
> changes happen
> on one or several lines of the picture).
> Either we have a good resolution/high intensity picture, and the
> intensity is
> dropping, or we have a poor resolution/low intensity picture, and
> the intensity
> is abruptly increasing. This seems to (although we are not sure yet
> about this
> part) increase when the confocal is on for a long time.
> Did somebody experience the same phenomenon with a similar system?
> The technician from Zeiss told us it could be a problem with one the
> acquisition cards. However, our end of the year budget does not
> allow us to
> change them. Is there any way we can try to at least minimize the
> problem?
> Thanks a lot.
> Regards
> Sandrine
|
|
|