Subject: | |
From: | |
Reply To: | |
Date: | Wed, 30 Apr 2003 09:07:30 -0500 |
Content-Type: | text/plain |
Parts/Attachments: |
|
|
Search the CONFOCAL archive at
http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
Greetings,
Cooincidently I discovered we had oil in our 63x oil objective yesterday
(we have an inverted platform). Has anyone attempted to use a vacum
evaporator to remove the oil from a compromised objective? Obviously
this won't repair the seal, but it may buy us a little time if we are
careful. Does anyone happen to know the vapor pressor of Cargille type
DF? Thanks in advance.
-Karl G.
Kenneth A. Orndorff wrote:
>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Confocal BB,
> Thanks everyone for the tips on cleaning. I would like to say that we do indeed use these good cleaning habits and we try to take excellent care of these very expensive elements of the system. It is really the more serious issue of what I believe is objective failure at the gasket that I wanted to address.
> Tom Gore is reporting exactly the type of problem that I think we should expect to be able to avoid with reasonable care and treatment of our objectives. I am especially concerned about objective failure because our core facility has just purchased a new inverted confocal system that will emphasize live cell work. Therefore, the inverted objective will be far more easily covered in oil. I expect that the higher 37C temp also places a much greater demand on the seals of the objectives. We have a water immersion objective for the long term studies of live cells. But other core facility users with fixed samples will also want to use the system when it is warmed, so the issue of oil running down the objective will still occur.
> Are microscopists to consider $5000.00 + objectives as disposable items? I would like to think not. This is my first experience with a gasket failure in nearly 25 years of working with microscopes. Thus, my first response was that something funny was going on. So I am just checking with with everyone to see if similar events are happening out there.
> I wonder if it might be possible for microscope companies to recertify the objectives for a much reduced cost than required when a total failure occurs. Possibly providing a loaner? I can dream. :-) Is it possible to remove aging but intact sealant and apply a new coating at a much reduced cost and much more quickly? Is it possible to avoid designs that have readily exposed sealing gaskets? Any thoughts that you have would be of interest to me. Thanks again to all.
>
>
>Ken
>
>Ken Orndorff
>Supervisor-Imaging
>H.C. Englert Cell Analysis Lab
>Dartmouth College/DMS/NCCC
>Borwell Building Room 338 West
>One Medical Center Drive
>Lebanon, NH 03756
>
>Phone 603-650-7661
>FAX 603-650-6130
>http://www.dartmouth.edu/~celllab
>
>
|
|
|