Andreas, can expound on the sentence below? I'm not sure what you mean
exactly.
Thank you!

"In strongly scattering samples like lymph nodes the signal decreases so
rapidly with depth that you hardly notice a difference in penetration
depth when adjusting the pre-chirp".

Brian D Armstrong PhD
Light Microscopy Core Manager
Beckman Research Institute
City of Hope
Dept of Neuroscience
1450 E Duarte Rd
Duarte, CA 91010
626-256-4673 x62872
http://www.cityofhope.org/research/support/Light-Microscopy-Digital-Imag
ing/Pages/default.aspx

-----Original Message-----
From: Confocal Microscopy List [mailto:[log in to unmask]]
On Behalf Of Andreas Bruckbauer
Sent: Thursday, November 18, 2010 11:09 PM
To: [log in to unmask]
Subject: Re: MaiTai DeepSee adjustments on multiphoton microscope

*****
To join, leave or search the confocal microscopy listserv, go to:
http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy
*****


 Hi Michael,

i tried a quantification at 800 nm excitation 450 - 500 micrometer into
fixed lymph nodes using the 25x Olympus objective. Over the range of the
deepsee compensation i found a change of a factor two in fluorescence.
One thing is that i could not set the compensation to zero with our
deepsee model  so i don't know what the fluorescence would be without
deepsee. Anyhow with a good estimate of the GDD of the system from the
deepsee motor position one can calculate how much this would broaden the
100 fs pulse out of the MaiTai and for our system this is about 220 fs.
so you would have to use 1.5x more laser power to get the same signal
without pre-chirp. 

In my experience you don't need to optimise for different samples as the
less than one mm depth into the sample is insignificant compared to the
thickness of the optics and GVD of an AOM. There is a slight difference
for different objectives though (more or less glass). My conclusion was
that for live samples the pre-chirp is definetly a benefit because you
can use lower laser power. In strongly scattering samples like lymph
nodes the signal decreases so rapidly with depth that you hardly notice
a difference in penetration depth when adjusting the pre-chirp. 

best wishes

Andreas


 

 

-----Original Message-----
From: Cameron Nowell <[log in to unmask]>
To: [log in to unmask]
Sent: Thu, 18 Nov 2010 21:58
Subject: Re: MaiTai DeepSee adjustments on multiphoton microscope


*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



Hi Michael,



We have the DeepSee on our MaiTai and have found that it doesn't do a

lot on our system. At extreme depths, say greater than 500um you can get

a small enhancement in signal strength and quality but it is only very

minor. It may give you a few extra microns penetration but thats about

it.



It really depends on how much the optics of your system are

dispersing/chirping the beam. If they are not altering it much then

pre-chirping it will not do a lot.



This is all based on qualitative assessment, i have not quantitated any

of this. Also we have recently received a new Olympus 25x NA1.05

objective specifically designed for MP imaging. I have not seen how this

performs yet with the DeepSee.





Cheers



Cam







Cameron J. Nowell

Microscopy Manager 

Centre for Advanced Microscopy 

Ludwig Institute for Cancer Research 

PO Box 2008 

Royal Melbourne Hospital 

Victoria, 3050 

AUSTRALIA 

Office: +61 3 9341 3155 

Mobile: +61422882700 

Fax: +61 3 9341 3104 

Facility Website







-----Original Message-----

From: Confocal Microscopy List [mailto:[log in to unmask]]

On Behalf Of Cammer, Michael

Sent: Friday, 19 November 2010 4:08 AM

To: [log in to unmask]

Subject: MaiTai DeepSee adjustments on multiphoton microscope



*****

To join, leave or search the confocal microscopy listserv, go to:

http://lists.umn.edu/cgi-bin/wa?A0=confocalmicroscopy

*****



Dear microscopists,



I'm planning to go through a regimen of  adjusting  the MaiTai DeepSee

on a variety of samples with different optics  to see what changes there

are in depth penetration (and, maybe, preservation of biological

activity).

  

Does anybody have experience with this?  If so, have you found that the

DeepSee provides a noticeable practical benefit over the MaiTai without

it?



Any comments welcome and please feel free to contact me offline at

[log in to unmask]



Thank you!



Sincerely,



Michael C.



------------------------------------------------------------

This email message, including any attachments, is for the sole use of

the intended recipient(s) and may contain information that is

proprietary, confidential, and exempt from disclosure under applicable

law. Any unauthorized review, use, disclosure, or distribution is

prohibited. If you have received this email in error please notify the

sender by return email and delete the original message. Please note, the

recipient should check this email and any attachments for the presence

of viruses. The organization accepts no liability for any damage caused

by any virus transmitted by this email.

=================================


 


---------------------------------------------------------------------
SECURITY/CONFIDENTIALITY WARNING:  
This message and any attachments are intended solely for the individual or entity to which they are addressed. This communication may contain information that is privileged, confidential, or exempt from disclosure under applicable law (e.g., personal health information, research data, financial information). Because this e-mail has been sent without encryption, individuals other than the intended recipient may be able to view the information, forward it to others or tamper with the information without the knowledge or consent of the sender. If you are not the intended recipient, or the employee or person responsible for delivering the message to the intended recipient, any dissemination, distribution or copying of the communication is strictly prohibited. If you received the communication in error, please notify the sender immediately by replying to this message and deleting the message and any accompanying files from your system. If, due to the security risks, you do not wish to receive further communications via e-mail, please reply to this message and inform the sender that you do not wish to receive further e-mail from the sender. 

---------------------------------------------------------------------

ATOM RSS1 RSS2