Dear Kathy,
the maximum resolution for optic microscopy should be 200 nm, according 
to standard physics laws: that means that any higher resolution will 
give you no more information on the sample.
As for useful formulas:
-       the Nyquist theorem for lateral resolution should be similar to 
X= 0,4*lambda/NA, where NA is numerical aperture of your objective and 
lambda the wavelength. It means that, if you’re working with 63x 
objective and 1,32 as NA, at 488nm, the lateral resolution is 148nm; 
dividing this number by 3, you obtain almost 49 nm. 49 is the raster 
distance you should set in your scan, in order to avoid loss of info 
from the specimen —> it does NOT mean that you will obtain resolution 
higher than 200nm, but only that you’ll obtain good data by scanning 
thrice the 148nm distance obtained before; in order to modify the 
raster distance you should change the electronic zoom  and the scan 
format.
At least, this is what I've found; if anyone else has nice data, I'll 
be  interested as you in reading more :)
Thank you and have a nice time.
AM




dott. Andrea Manazza
Dipartimento di Oncologia, Università di Torino
CeRMS, ospedale Molinette
Via Santena, 7 - 10126 Torino, Italia
tel: +39-11-6706522
lab: +39-11-6336859
fax: +39-11-6336887

----- Original Message -----
From: Kathy Spencer <[log in to unmask]>
Date: Wednesday, November 27, 2002 9:32 pm
Subject: confocal resolution

> Search the CONFOCAL archive at
> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
> 
> Hello All!
>         Basic generic question....what is the limit of resolution 
> on a
> confocal? I'm using an Olympus Fluoview confocal, 60x 1.4NA 
> objective, oil
> (1.516), pinhole automatically set for 1 Airy disk. For 488nm 
> laser, I
> calculate 174nm using resolution=wavelength/2NA. But can the confocal
> resolve more than that? What is a good formula?
>         Thanks!
>         Kathy
> 
> 
> 
> 
> 
> 
> 
> 
> Kathy Spencer
> The Scripps Research Institute
> 10550 N. Torrey Pines Road
> ICND 202
> La Jolla, CA 92037
> 858-784-8437
>