Dear Karl,
Guy is correct here. I've checked our 4 PMTs against bluish through deep
red fluorescence, and PMT1 is definitely most sensitive in the blue, PMT2
in green-yellow, PMT3 in orange-red and PMT4 in red through long red
Another quirk to be aware of is that the dichroic filters let through
different % of laser light. So the intensity of the 488 line coming
through RSP500, DD488/543 and TD488/543/633 is different for each.
cheers,
Rosemary
>Search the CONFOCAL archive at
>http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>
>Quoting Karl Garsha <[log in to unmask]>:
>
>> Search the CONFOCAL archive at
>> http://listserv.acsu.buffalo.edu/cgi-bin/wa?S1=confocal
>>
>> Greetings,
>> Has anybody determined a benchmark for evaluating PMT sensitivity on the
>> SP-2 platform? I've recently discovered that our PMT 1 is showing
>> roughly 30% of the sensitivity of PMT 3 and PMT 2 is showing 65% of the
>> sensitivity of PMT 3 (at either end of the spectrum). This is a little
>> alarming and I would like to determine whether PMT 3 is in fact
>> performing well or if the sensitivity is down for it as well. Thanks in
>> advance for any suggestions.
>
>Depending on your system they may well be different types of PMT -
>I think our system has 1 bi-alkali and the others multi-alkali. So
>check that first. Generally one would expect that more blue-sensitivity
>is desirable on PMT1 and more red on PMT3. The other thing to remember
>is that voltage response may vary and a more accurate control is
>obtainable signal to noise ratio.
>
> Guy
>
>--